Comment AG4100 SNF 2009-04-23 08:45:38: Max run time data

eenriquez at snf.stanford.edu eenriquez at snf.stanford.edu
Thu Apr 23 08:45:39 PDT 2009


Ran a test where I manually set the lamp intensity to obtain a temperature while monitoring the top plate temperature.  The purpose of this test is to determine how long you could run at a particular temperature before the top plate over temp switch trips (89C - 90C)
@ 650 C
	150 sec - 52C
	300 sec - 66C
	450 sec - 73C
	600 sec - 76C
	750 sec - 79C
@ 750 C
	150 sec - 65C
	300 sec - 82C
	(430) sec - 89C     Lamps shut-off at 430 sec