From jackson at snf.stanford.edu Wed Mar 4 01:04:23 2009 From: jackson at snf.stanford.edu (jackson at snf.stanford.edu) Date: Wed, 4 Mar 2009 01:04:23 -0800 Subject: Problem AG4108 SNF 2009-03-04 01:04:23: overtemp top plate on 1st wafer Message-ID: running only 1050 for 30 sec, program has been proven to be ok before From rshyam at snf.stanford.edu Fri Mar 6 04:49:08 2009 From: rshyam at snf.stanford.edu (rshyam at snf.stanford.edu) Date: Fri, 6 Mar 2009 04:49:08 -0800 Subject: Problem AG4108 SNF 2009-03-06 04:49:08: overtemp hot plate Message-ID: tried to run RTA1000 1 min From emyers at snf.stanford.edu Fri Mar 6 14:53:57 2009 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Fri, 6 Mar 2009 14:53:57 -0800 Subject: Comment AG4108 SNF 2009-03-06 14:53:57: Could Not Resolve Message-ID: I was not able to clear the top plate problem. I reset 2 of the 3 switches and Elmer also attempted to clear the faults. John Shott or Cesar may know where the last remaining interlock is located. From shott at snf.stanford.edu Mon Mar 23 09:15:33 2009 From: shott at snf.stanford.edu (shott at snf.stanford.edu) Date: Mon, 23 Mar 2009 09:15:33 -0700 Subject: Shutdown AG4108 SNF 2009-03-23 09:15:33: Adding auto reset overtemp switches ... Message-ID: Should be back up by 11 a.m. John From shott at snf.stanford.edu Mon Mar 23 13:18:25 2009 From: shott at snf.stanford.edu (shott at snf.stanford.edu) Date: Mon, 23 Mar 2009 13:18:25 -0700 Subject: Shutdown AG4108 SNF 2009-03-23 09:15:33: Adding auto reset overtemp switches ... Message-ID: Added the second auto-reset over temp switch and applied adhesive type K thermocouples near each one. Still need to attach the "feedthrought" and attache the thermocouple plug. System is OK to use ... John