oxidation recipe
Raymond Woo
rwoo at stanford.edu
Fri Aug 18 16:37:56 PDT 2006
We ran an oxidation test this afternoon with 4 wafers using the recipe
that Robin suggested below. The actual ramp rate was a bit low (~60
deg/sec) but that can probably be fixed with an optimized recipe. The
ellipsometer measured oxide thicknesses varied between 102A and 117A.
From TSUPREM simulations this corresponds to a worst case temperature
error of between 5 and 10 deg. I've kept the individual wafer
printouts, so if anyone would like the more detailed results just let me
know.
First Wafer: avg=108 std dev=2.8
Second Wafer: avg=102 std dev=3.9
Third Wafer: avg=115 std dev=3.0
Fourth Wafer: avg=117 std dev=3.8
- Ray
Robin King wrote:
> For anyone wanting to check the tempurature calibration on the 4108
> or 4100 so they can run wafers, it's easy to diffusion clean some
> lightly doped test wafers and run this recipe. Be sure to run a few
> cleaned dummy wafers ahead of the real test wafers.
> AG field service sent a copy of the conditions they use: O2, 1150C,
> 30 seconds, ramp at 100C/sec. Result should be 120A or close using
> the ellipsometer program 11.
> It would be good to start doing this again on a regular basis, such
> as during training etc.
>
>
>
>
>
> __________________________________________________
> Do You Yahoo!?
> Tired of spam? Yahoo! Mail has the best spam protection around
> http://mail.yahoo.com