Problem amtetcher SNF 2007-05-09 00:45:32: turbo pump cannot pump down to base
eenriquez at snf.stanford.edu
eenriquez at snf.stanford.edu
Wed May 9 08:47:35 PDT 2007
Hot loop water for the bell jar is only at 27 C, setpoint is 35C. Need to order a new heater. Ran a pumping speed vs. time bell jar is open. When the bell jar has been open for 2 min., the pump down time is 2.9 min. When the bell jar open time = 10 min., the pump down time = 5.8 min.