Problem asml SNF 2008-04-04 08:33:37: Carrier inspection problem
gsosa at snf.stanford.edu
gsosa at snf.stanford.edu
Fri Apr 4 08:45:12 PDT 2008
System event log indicated low intensity during carrier inspection. Was able to reproduce the problem several times. Found that the hinges for port #4 cover were loose causing misposition when the cover was closed. When the cover was mis-positioned enough the sensor to activate the inspecion laser was not satisfied and the inspection would fail. Tightened up hinges on port # 4 and port #3. Retested carrier inspection several times opening andclosing cover inbetween the tests. All OK now.