Cascade manipulator broken
mihirt at stanford.edu
mihirt at stanford.edu
Wed May 16 19:25:46 PDT 2007
Quoting "Deji A." <dejia at stanford.edu>:
> What is the root cause of the broken manipulator so we can avoid
> this problem in the future. What needs to be done to avoid breaking
> the manipulator?
Good question. Sorry I wasn't more clear.
In this case, we determined that the issue is a wear-and-tear
problem. Older users might recall that the Cascade used to have a
shared set of probe tips - but due to frequent damage, each user was
required to provide his/her own probe tips. So each user
removes/inserts probe tips during their session. It is this continuous
action that wears down the contacts inside the manipulator.
There really isn't a way around this. We can return to a common
set of probe tips to minimize wear, but that will result in damaged
probe tips from careless users. Around and around we go...
One thing that would help - from all users - is making sure that
the stage and manipulators are clean *before and after* your session.
Use wipes with a little bit of IPA to clean the surface and bottoms of
the manipulators. This prevents vacuum problems.
- Mihir