Comment drytek2 SNF 2009-09-10 17:52:35: Update ... still manometer problems ...
shott at snf.stanford.edu
shott at snf.stanford.edu
Fri Sep 11 23:08:32 PDT 2009
We have replaced the pump and the capacitance manometer.
The manometer is a newer MKS Model 627 1 Torr head rather than the older Model 227 that was in service previously. We have a custom wiring harness to adapte the existing controller to the new head.
This head should produce 0-10 output for a pressure of 0-1 torr. In other words, 150 mTorr should produce 1.5 V output.
For a time, we though that we had a blower problem becauase the "normal" nitrogen flow conditions was reading (it appeared) about a 10X higher pressure than normal .... that is 3400 rather than 340 mTorr. However, a 1 Torr head CAN'T read 3400 mTorr (because that would be an output of 34 V which can't happen for a head supplied by +/- 15 V.
As a result, I believe that this was actually reading 340.0 Torr and we had failed to set the decimal point.
Also, because this is a 1 Torr head (Drytek1 uses a 10 Torr head), we need the inputy voltage on teh Type 252A pressure controller to be set to
10V input rather than 1 V input.
We will confirm these things on Monday with test wafers and with more careful pressure and voltage measurements.
Please process with caution until that determination has been made.
Thanks,
John
More information about the drytek2-pcs
mailing list