Tentative clear of shutdown ... but use with extreme caution!!!
Raja Jain
jrjain at stanford.edu
Sun Sep 13 17:04:38 PDT 2009
Fellow drytek2 users,
For those of you customarily using drytek2 for descum, gasonics may be a
reasonable backup based on the tests below (3um 220-3 on bare Si). (Thanks
to Jim McVittie for the recipe suggestions)
Test #1: Gasonics 0A
Resist removed from wafer center: 150A
Test #2: Gasonics 0D
Resist removed from wafer center: 4084A (likely due to strong heating
effects)
Resist thicknesses were measured using the litho nanspec.
Regards,
Raj
-----Original Message-----
From: John Shott [mailto:shott at stanford.edu]
Sent: Friday, September 11, 2009 11:59 PM
To: drytek2 at snf.stanford.edu
Subject: Tentative clear of shutdown ... but use with extreme caution!!!
SNF Lab Members:
I have just cleared the shutdown on drytek2 .... but want to make sure
that you all read this very carefully before proceeding with any
critical etching.
As you know, this machine has been experiencing pump down and vacuum
readout problems. We have now replaced the main vacuum pump as well as
the Capacitance Manometer with a brand new 1 Torr head. The other
dryteks (1 and 4) use a 10 Torr head.
This difference is why I am asking you to proceed with caution in any
etching that you do with drytek2 this weekend. While I believe that we
are reading and controlling to a proper pressure .... that is, when it
reads 150.0 it is actually 150 mTorr, there is a chance that that is not
correct.
Why the confusion? The 10 Torr heads on dryteks 1 and 4 produce a 0-10
V output signal when the chamber pressure varies from 0 - 10 Torr ... so
that a 150 mTorr vacuum should produce 0.150 V. The 1 Torr head on
drytek2, however, produces 0 - 10 V output as the pressure varies from
0-1 torr ....so that 150 mTorr vacuum produces 1.50 V. I believe that I
have properly set the pressure controller on drytek2 to accommodate this
10X more sensitive capacitance monitor.
In any event, I THINK that everything is set up properly .... but we
have not had time to yet run test wafers or make corroborating pressure
measurements. As a result, if you have a critical etch for drytek2, I
strongly suggest that you wait until Elmer and Nancy can perform
additional machine tests and run test wafers, respectively, to make sure
that I am not in error. If, on the other hand, you have a non-critical
usage, such as a descum, you may be willing to run some wafers this
weekend, I'd still recommend running test wafers just to be sure ....
but I know that there may be some pent up demand for this tool, so I
wanted to clear the shutdown so that you could use it .... but also warn
you that while I believe we are controlling pressure correctly, that we
may still have a pressure readout problem.
Thanks for your continued support,
John
More information about the drytek2
mailing list