Shutdown ellipsomter SNF 2006-10-09 11:25:56: microscope lamp is not working anymore
emyers at snf.stanford.edu
emyers at snf.stanford.edu
Wed Oct 25 10:58:42 PDT 2006
Ted replaced microscope and ACT bulbs. I adjusted stage height (which was off) and ran Uli's calibration wafer. Results thickness 1045 vs. target of 1045, index of 1.472 vs target of 1.47.