From gsosa at snf.stanford.edu Wed Jul 7 13:50:40 2010 From: gsosa at snf.stanford.edu (gsosa at snf.stanford.edu) Date: Wed, 7 Jul 2010 13:50:40 -0700 Subject: Comment evalign SNF 2010-07-07 13:50:40: Intensity Check 7/7/10 Message-ID: Intensity Check: Intensity Ave: 9.73 Mw/cm2 Uniformity: 1.7% From gsosa at snf.stanford.edu Fri Jul 9 09:21:04 2010 From: gsosa at snf.stanford.edu (gsosa at snf.stanford.edu) Date: Fri, 9 Jul 2010 09:21:04 -0700 Subject: Comment evalign SNF 2010-07-09 09:21:03: Scheduled lamp change 7/9/10 Message-ID: Replaced UV lamp and adjusted lamp position for max intensity. Adjusted power supply for 15.0 mw/cm2 intensity. Intensity Check: Intensity Average: 15.13 Uniformity: 1.52% From gsosa at snf.stanford.edu Mon Jul 12 09:13:45 2010 From: gsosa at snf.stanford.edu (gsosa at snf.stanford.edu) Date: Mon, 12 Jul 2010 09:13:45 -0700 Subject: Comment evalign SNF 2010-07-12 09:13:45: Intensity Check 7/12 Message-ID: Intensity Check 7/12 Intensity Start: 13.60 mW/cm2 Intensity Calibrated To: 15.07 mW/cm2 Uniformity: 1.33% From vilanova at snf.stanford.edu Thu Jul 15 08:32:02 2010 From: vilanova at snf.stanford.edu (vilanova at snf.stanford.edu) Date: Thu, 15 Jul 2010 08:32:02 -0700 Subject: Comment evalign SNF 2010-07-15 08:32:01: weekly Intensity + Uniformity results Message-ID: Intensity Average = 13.77Mw/Cm2 Uniformity = 1.57% From gsosa at snf.stanford.edu Wed Jul 21 14:44:03 2010 From: gsosa at snf.stanford.edu (gsosa at snf.stanford.edu) Date: Wed, 21 Jul 2010 14:44:03 -0700 Subject: Comment evalign SNF 2010-07-21 14:44:02: Intensity Check 7/21 Message-ID: Intensity Check Intensity Average: 12.07 mW/cm2 Uniformity: 0.66%