Shutdown headway2 SNF 2007-07-13 09:42:02: Intermittent odors observed

gsosa at gsosa at
Fri Jul 13 14:20:41 PDT 2007

Removed the coater and found exhaust hose to bottom of coater disconnected. Re-installed the exhaust hose and secured with a hose clamp. THere is now exhaust flow inside of the coater bowl. Verified exhaust inside plenum- OK. Checked airflowat front edge of wet deck near breathing zone. Airflow was 100 to 150 Ft/min using a hotwire annemometer. System ready for use.