From jhaydon at snf.stanford.edu Wed Jul 1 15:23:48 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Wed, 1 Jul 2009 15:23:48 -0700 Subject: Shutdown innotec SNF 2009-07-01 15:23:47: bad signal from VSC Message-ID: From jhaydon at snf.stanford.edu Wed Jul 1 15:38:16 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Wed, 1 Jul 2009 15:38:16 -0700 Subject: Shutdown innotec SNF 2009-07-01 15:23:47: bad signal from VSC Message-ID: miss read tag on cable adjusted switch From mrlin at snf.stanford.edu Wed Jul 8 18:47:27 2009 From: mrlin at snf.stanford.edu (mrlin at snf.stanford.edu) Date: Wed, 8 Jul 2009 18:47:27 -0700 Subject: Problem innotec SNF 2009-07-08 18:47:26: no high voltage Message-ID: went through standard procedure, but when it says " to check for 10kV", the reading is still ~0kV. From gdiankov at snf.stanford.edu Wed Jul 8 21:37:21 2009 From: gdiankov at snf.stanford.edu (gdiankov at snf.stanford.edu) Date: Wed, 8 Jul 2009 21:37:21 -0700 Subject: Comment innotec SNF 2009-07-08 21:37:20: HV back on. Vaccuumed thoroughly. Message-ID: From jperez at snf.stanford.edu Thu Jul 9 14:11:18 2009 From: jperez at snf.stanford.edu (jperez at snf.stanford.edu) Date: Thu, 9 Jul 2009 14:11:18 -0700 Subject: Shutdown innotec SNF 2009-07-09 14:11:17: Power Supply Problem Message-ID: All standard lights are coming on for the power supply except for the DC kilovolts. The base pressure isn't pumping below 1x10 -6 range. At one point the Temperature of the cryo was at 18 / 21K. It is back to 14/ 15 From cbaxter at snf.stanford.edu Thu Jul 9 20:45:38 2009 From: cbaxter at snf.stanford.edu (cbaxter at snf.stanford.edu) Date: Thu, 9 Jul 2009 20:45:38 -0700 Subject: Shutdown innotec SNF 2009-07-09 14:11:17: Power Supply Problem Message-ID: Vented chamber and removed the shielding , removed e-gun clean hv contacts and re-assemble. Run dep with al source 1kA. Pumped 7.7e-7 in 15 minutes. Thank s to EricP. From cbaxter at snf.stanford.edu Thu Jul 9 20:45:56 2009 From: cbaxter at snf.stanford.edu (cbaxter at snf.stanford.edu) Date: Thu, 9 Jul 2009 20:45:56 -0700 Subject: Comment innotec SNF 2009-07-08 21:37:20: HV back on. Vaccuumed thoroughly. Message-ID: From cbaxter at snf.stanford.edu Thu Jul 9 20:46:24 2009 From: cbaxter at snf.stanford.edu (cbaxter at snf.stanford.edu) Date: Thu, 9 Jul 2009 20:46:24 -0700 Subject: Problem innotec SNF 2009-07-08 18:47:26: no high voltage Message-ID: Problem fixed see shutdown report.. From cbaxter at snf.stanford.edu Thu Jul 9 20:47:13 2009 From: cbaxter at snf.stanford.edu (cbaxter at snf.stanford.edu) Date: Thu, 9 Jul 2009 20:47:13 -0700 Subject: Comment innotec SNF 2009-07-09 20:47:12: Test dep ok Message-ID: Al source still loaded, ok to remove.. From gdiankov at snf.stanford.edu Sat Jul 11 20:16:52 2009 From: gdiankov at snf.stanford.edu (gdiankov at snf.stanford.edu) Date: Sat, 11 Jul 2009 20:16:52 -0700 Subject: Problem innotec SNF 2009-07-11 20:16:51: No beam seen; aborted. Message-ID: From jennyhu at snf.stanford.edu Sun Jul 12 09:58:26 2009 From: jennyhu at snf.stanford.edu (jennyhu at snf.stanford.edu) Date: Sun, 12 Jul 2009 09:58:26 -0700 Subject: Problem innotec SNF 2009-07-12 09:58:25: Can't see beam through view port Message-ID: I can't see the beam through the view port. The beam is definitely on though, because the DC current increases, and I can see a glow. Film is also deposited. From jhaydon at snf.stanford.edu Mon Jul 13 08:41:30 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 13 Jul 2009 08:41:30 -0700 Subject: Shutdown innotec SNF 2009-07-13 08:41:29: down for regen Message-ID: should be back up around 2:00 PM From jperez at snf.stanford.edu Mon Jul 13 14:36:27 2009 From: jperez at snf.stanford.edu (jperez at snf.stanford.edu) Date: Mon, 13 Jul 2009 14:36:27 -0700 Subject: Shutdown innotec SNF 2009-07-13 14:36:26: Re-gen stuck and didn't continue, estimated up time 5 PM Message-ID: From jhaydon at snf.stanford.edu Mon Jul 13 16:53:27 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 13 Jul 2009 16:53:27 -0700 Subject: Shutdown innotec SNF 2009-07-13 14:36:26: Re-gen stuck and didn't continue, estimated up time 5 PM Message-ID: From jhaydon at snf.stanford.edu Mon Jul 13 16:54:19 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 13 Jul 2009 16:54:19 -0700 Subject: Shutdown innotec SNF 2009-07-13 08:41:29: down for regen Message-ID: ok to use when cryo temperature is at 16K should be about 6:30 PM From jperez at snf.stanford.edu Tue Jul 14 07:05:00 2009 From: jperez at snf.stanford.edu (jperez at snf.stanford.edu) Date: Tue, 14 Jul 2009 07:05:00 -0700 Subject: Problem innotec SNF 2009-07-12 09:58:25: Can't see beam through view port Message-ID: Not sure why User couldn't see through polarize window. We were able to see. From jperez at snf.stanford.edu Tue Jul 14 07:07:04 2009 From: jperez at snf.stanford.edu (jperez at snf.stanford.edu) Date: Tue, 14 Jul 2009 07:07:04 -0700 Subject: Problem innotec SNF 2009-07-11 20:16:51: No beam seen; aborted. Message-ID: Film knob may have been turned in the wrong direction, but when we took a look at it, no problem. From jperez at snf.stanford.edu Tue Jul 14 07:07:28 2009 From: jperez at snf.stanford.edu (jperez at snf.stanford.edu) Date: Tue, 14 Jul 2009 07:07:28 -0700 Subject: Comment innotec SNF 2009-07-09 20:47:12: Test dep ok Message-ID: Thanks once again! From ghyrn at snf.stanford.edu Tue Jul 21 18:31:45 2009 From: ghyrn at snf.stanford.edu (ghyrn at snf.stanford.edu) Date: Tue, 21 Jul 2009 18:31:45 -0700 Subject: Comment innotec SNF 2009-07-21 18:31:45: Ni source gone, Shutter replaced Message-ID: From emyers at snf.stanford.edu Wed Jul 22 12:42:02 2009 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Wed, 22 Jul 2009 12:42:02 -0700 Subject: Comment innotec SNF 2009-07-21 18:31:45: Ni source gone, Shutter replaced Message-ID: Replaced Ni source. Did not find damaged Ni source. From emyers at snf.stanford.edu Wed Jul 22 12:42:34 2009 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Wed, 22 Jul 2009 12:42:34 -0700 Subject: Comment innotec SNF 2009-07-22 12:42:34: New Ni and Cu Sources Message-ID: Placed new Ni and Cu sources in source box. From kenney at snf.stanford.edu Sat Jul 25 00:26:13 2009 From: kenney at snf.stanford.edu (kenney at snf.stanford.edu) Date: Sat, 25 Jul 2009 00:26:13 -0700 Subject: Problem innotec SNF 2009-07-25 00:26:12: WON'T PUMP DOWN Message-ID: After pushing "Pump Only" it sounds like it is sucking air and the pressure doesn't drop. From kenney at snf.stanford.edu Sat Jul 25 00:37:16 2009 From: kenney at snf.stanford.edu (kenney at snf.stanford.edu) Date: Sat, 25 Jul 2009 00:37:16 -0700 Subject: Problem innotec SNF 2009-07-25 00:37:16: Pumping fine now Message-ID: Everything is fine Machine is working properly From filip at snf.stanford.edu Sun Jul 26 10:23:49 2009 From: filip at snf.stanford.edu (filip at snf.stanford.edu) Date: Sun, 26 Jul 2009 10:23:49 -0700 Subject: Comment innotec SNF 2009-07-26 10:23:48: planetary motion not smooth Message-ID: when checked before pumpdown, it was OK. During deposition it looks like it stutters as it rotates. From jhaydon at snf.stanford.edu Mon Jul 27 07:29:31 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 27 Jul 2009 07:29:31 -0700 Subject: Problem innotec SNF 2009-07-25 00:26:12: WON'T PUMP DOWN Message-ID: problem fixed by user From jhaydon at snf.stanford.edu Mon Jul 27 07:29:35 2009 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 27 Jul 2009 07:29:35 -0700 Subject: Problem innotec SNF 2009-07-25 00:37:16: Pumping fine now Message-ID: From helencxy at snf.stanford.edu Fri Jul 31 19:33:55 2009 From: helencxy at snf.stanford.edu (helencxy at snf.stanford.edu) Date: Fri, 31 Jul 2009 19:33:55 -0700 Subject: Comment innotec SNF 2009-07-31 19:33:55: need more gold pallets Message-ID: Gold target is still low, and all pallets are gone.