From emyers at snf.stanford.edu Fri Nov 4 15:41:06 2011 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Fri, 4 Nov 2011 15:41:06 -0700 Subject: Shutdown intlvac_evap SNF 2011-11-04 15:41:06: Cryo not cooling down Message-ID: I watched the regen process for the last 2 hours. I sucesfully went through the warm-up, pump down and rate of rise. Now an hour later it has not cooled down at all. The compressor is running. From emyers at snf.stanford.edu Tue Nov 15 08:59:47 2011 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Tue, 15 Nov 2011 08:59:47 -0800 Subject: Shutdown intlvac_evap SNF 2011-11-04 15:41:06: Cryo not cooling down Message-ID: Jim replaced the controller and the system is now under high vac From emyers at snf.stanford.edu Wed Nov 16 13:42:50 2011 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Wed, 16 Nov 2011 13:42:50 -0800 Subject: Shutdown intlvac_evap SNF 2011-11-16 13:42:50: Source out for repair Message-ID: The ebeam source is back at the OEM for re-gausing and pole piece set-up From jhaydon at snf.stanford.edu Mon Nov 21 15:26:33 2011 From: jhaydon at snf.stanford.edu (jhaydon at snf.stanford.edu) Date: Mon, 21 Nov 2011 15:26:33 -0800 Subject: Shutdown intlvac_evap SNF 2011-11-16 13:42:50: Source out for repair Message-ID: source reinstalled From emyers at snf.stanford.edu Wed Nov 23 07:44:14 2011 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Wed, 23 Nov 2011 07:44:14 -0800 Subject: Comment intlvac_evap SNF 2011-04-20 16:22:25: Restarted after EMO Message-ID: archived