Shutdown karlsuss SNF 2009-07-29 07:36:44: Shutter repair

gsosa at snf.stanford.edu gsosa at snf.stanford.edu
Wed Jul 29 10:58:10 PDT 2009


Replaced both air cylinders for the exposure shutters. Tested and adjusted pneumatics for proper timing. Replaced the UV lamp. Adjusted lamp position for max intensity . Calibrated lamp power supply to 15 mw/cm2. Uniformity at 1.6%. System ready to use.