Shutdown karlsuss SNF 2011-06-08 08:52:31: report of misalignment
mahnaz at snf.stanford.edu
mahnaz at snf.stanford.edu
Wed Jun 8 10:04:28 PDT 2011
Mario adjusted the screws on the stage( they were totally loose).
Y macrometer was totally out, was adjusted.
Mario checked the lamp uniformity is within spec. 15mw. and 2%
I tested the system with 3612 1 um , 1.1 seconds of exposure and
resolved 2 um line and space.
cleaned the system off all the tapes !!!!!