Comment laurell-R SNF 2011-08-15 20:51:50: Vacuum pressure low

gsosa at gsosa at
Tue Aug 16 16:07:27 PDT 2011

Vacuum at spindle only 13.8" Hg. Not enough to satisfy the vacuum sensor. Verified incoming supply vacuum- OK at 25" Hg. Verified all pneumatic lines and solinoid- All OK. Isolated loss of spindle vacuum to shaft seal. Replaced the shaft seal. Now vacum at spindle is 21.3" Hg with  a wafer and chuck. Tested operation with standard program. All OK now.

More information about the laurell-R-pcs mailing list