Comment metalica 2003-11-22 15:12:35: main chamber vacuum gauge
zia at snf.stanford.edu
zia at snf.stanford.edu
Sat Nov 22 15:12:35 PST 2003
the main vacuum gauge seems to be misreading. after a fifteen minute roughing period, the reading was only at ~200 microns. After checking for open lines, I decided to try the cryopump. Immediately the reading dropped, but did not go below 50 microns. Using the ion gauge though, the chamber was at 1e-5 T. I let the chamber pump down for two hours to reach a base pressure of 6e-7 T, so my guess is that there is just something wrong with the main gauge.....
thanks,
rashid.