Problem micromanipulator6000 SNF 2012-10-15 22:05:40: cannot make measurement

shott at snf.stanford.edu shott at snf.stanford.edu
Sun Dec 23 09:09:24 PST 2012


I believe that Kai has found the source of this problem.  An open circuit in one of the triaxial probe connections prevents current from flowing.  As a result, if current is specified in that channel and compliance is greater than 42V, the interlock circuit will be reached as the voltage increases in a (futile) attempt to force current to flow.
Kai has labeled and set aside the offending probe assembly.
Thanks,
John