[micromanipulator6000-pcs] Shutdown micromanipulator6000 SNF 2013-04-23 16:12:14: Working on recovering and reinstalling EasyEXPERT

shott at stanford.edu shott at stanford.edu
Wed Apr 24 16:14:08 PDT 2013


Cleared EasyEXPERT software starting problem.  The messsge window: "Message ID: 103002" that says:
USB488 device was not found
Alias: AgB1500A
Reason: Our of resources
Indicated that the internal GP-IB controller is not available or not working.
A "lower level" test of this is to stop EasyEXPERT, and fire up "Programs --> Agillent IO Library Suite --> Agilent Connection Expert
In the Agilent ConnectionExpert, clicking "Refresh All" should try to connect to all hardware (and is a lower-level connection tool than EasyEXPERT.  The equipment Hierarchy should include:
B1500A
   COM1
   COM2
   GPIB0
   LAN
   USB0
      B1500A
         agb1500a
         usbDevice1
(This above hierarchy is likely to get scrambled ...)
Note: I also upgraded the firmware and EasyExpert software from version A.04.11.2010.1201 to A.05.03.2013.0128.
John