From shott at stanford.edu Wed Dec 5 13:05:14 2012 From: shott at stanford.edu (John Shott) Date: Wed, 05 Dec 2012 13:05:14 -0800 Subject: Measurements above 42 volts ... Message-ID: <50BFB70A.1050208@stanford.edu> Micromanipulator 6000 Community: Several of you have noticed that the Agilent B1500 limits measurements to +/- 42 volts even though the SMUs are capable of supplying up to +/- 100 volts. There is an interlock circuit on the back of the B1500 that is supposed to be connected to switches that prevent voltages higher than 42 volts from being applied unless protective enclosures are in place. I've learned that most other places using a B1500 have simply shorted out the interlock switch at the back of the unit. We have installed a pair of magnetic proximity switches in the probe station enclosure so that the interlock will be happy (that is the switches are closed) when the copper cover on the probe station is fully down. As a result those of you needing to apply voltages greater than 42 volts can no do so if you simply fully close the cover of the probe station enclosure. Please let me know if you have any questions, John