From mcvittie at snf.stanford.edu Thu Mar 28 12:06:23 2002 From: mcvittie at snf.stanford.edu (Jim McVittie) Date: Thu, 28 Mar 2002 12:06:23 -0800 Subject: Arcing Problem Message-ID: <3CA377BF.7F257BC0@snf.stanford.edu> P5000 Users, This week we had to replace some expensive parts in the oxide etch chamber because they were damaged from arcing. After discussing the problem in AMAT, we concluded that the problem was most likely from running the chamber at too high of a bias voltage. It was estimated that the bias voltage may have been as high as 1600V. As you know, the one does not directly set the bias voltage in a process recipe. Instead, the bias is developed from the RF current with a resulting bias value which depends on the pressure, power, B field and gas mixture. With a typical power setting of 500 to 600W, the most important parameter determining bias is the pressure. To avoid arcing in the future, Applied is recommending that we prohibit any process with pressures below 60 mT. The exception is the breakthrough step in the Al etch process which uses 30 mT at 300W for 15s to go through the Al2O3 layer. We will continue to let users modify the standard etch process on the P5000, but I need approve all proposed process modifications before hand. For your interest the process that lead to the arcing problem was: O2=20, He=20, Pressure=20mT, Power=600W, B=40G, Time=1000s Jim McVittie From mcvittie at snf.stanford.edu Thu Mar 28 14:29:36 2002 From: mcvittie at snf.stanford.edu (Jim McVittie) Date: Thu, 28 Mar 2002 14:29:36 -0800 Subject: P5000 Logbook References: <3CA377BF.7F257BC0@snf.stanford.edu> Message-ID: <3CA39950.222DBE28@snf.stanford.edu> P5000Users, In trying to understand the recent arcing problem, I found that some users are not fill out the log books. When problems occur, it is extremely important that we have a good log of the recent history of the chamber. In particular, I want to know # of wafers run, total rf on-time, any process changes for a standard process, bias voltages and anything unusual that occurred during the run. If we continue to have a problem with getting the logs fill out, some delinquent users may be removed from the P5000 qualified user list for a week or two. I expect, we can get compliance without having to go that far. By the way, there is still some debate going on about whether the 20 mT/600W process was the cause for the arcing damage. Thanks, Jim McVittie