Problem pquest SNF 2009-08-08 08:48:26: err6 on heater panel
jim kruger
jimkruger at yahoo.com
Sat Aug 8 09:27:52 PDT 2009
I saw this last night. You can reset the heater controller by holding in both "arrows" for a count of 10, then cycling through "Menu" until it is reading "temperature" again. I am not sure, but if you set the controller for "0 C", then watch the "L1" (heating) indicator to make sure it stays off, you can set the chiller for a few degrees below desired chuck temperature as we did before. As soon as the RF comes on, the Err6 reappears, but if the "L1" indicator stays off, I believe the chuck will stay at the chiller temperature.
Voltage vs. RF power is still low, I increased RF power but still did not get as much voltage as expected. My process worked, but it is not very sensitive to voltage.
I would not commit any really important sample.
jim
--- On Sat, 8/8/09, krivoire at snf.stanford.edu <krivoire at snf.stanford.edu> wrote:
> From: krivoire at snf.stanford.edu <krivoire at snf.stanford.edu>
> Subject: Problem pquest SNF 2009-08-08 08:48:26: err6 on heater panel
> To: pquest-pcs at snf.stanford.edu
> Date: Saturday, August 8, 2009, 8:48 AM
> heater panel displays err6 and
> temperature reads 60 despite being set for 10.
>
>