Chlorine at low flows ....
John Shott
shott at stanford.edu
Thu May 29 15:51:14 PDT 2008
Raith community:
This morning, Elmer did a nice job of cleanly re-installing the local
chlorine pressure regulator at the back of the machine so that it can
control the chlorine pressure to the pquest without affecting pressure
to either the lampoly or the p5000 etch. I believe that he has it set
at 8 PSI which we found allowed him to control flows to as low as 0.6
SCCM and yet still get full flows in excess of 40 SCCM. In the process
of doing this, however, he also found that the zero on the flowmeter was
off by a significant amount .... 1.1 SCCM as I recall.
I know that some of you run a recipe that has chlorine flows in the 1.5
- 2.5 SCCM range. As a result, particularly if your use one of these
low chlorine flow recipes, I suggest that you do a test run to make sure
that you are getting the etch results that you expect.
Secondly, we have received reports that the matching network (during
auto match, I think) may be misbehaving. At this point, however, it
does not seem to be failing consistently. Accordingly, it would be of
great help to is if you could:
1. Keep an eye on the performance of the matching network during your
etches.
2. If you see problems, please let us know what you observed and which
recipe and RF power levels you found to be problematics. This will help
us to more quickly resolve the problems.
Thanks,
John