From mihuhou at stanford.edu Mon Jun 1 22:55:43 2009 From: mihuhou at stanford.edu (Ying Chen) Date: Mon, 1 Jun 2009 22:55:43 -0700 (PDT) Subject: pquest released 6/2, 9:00am-12:00pm and more In-Reply-To: <4d36fb940904300837i7d69e3f0t435051fe7d01a922@mail.gmail.com> Message-ID: <1593653135.1560871243922143918.JavaMail.root@zm06.stanford.edu> Hi, Pquest users, I found the DC reading of pquest is 50~60% higher than previous when exactly the same recipe was used. Old Oxygen3.RCP has DC voltage of 170V, so I changed the RF to 50W (from 100W) to lower the DC reading to 113V. Ying From mihuhou at stanford.edu Thu Jun 4 18:53:36 2009 From: mihuhou at stanford.edu (Ying Chen) Date: Thu, 4 Jun 2009 18:53:36 -0700 (PDT) Subject: pquest problem report/pump might have been turned off. In-Reply-To: <4d36fb940904300837i7d69e3f0t435051fe7d01a922@mail.gmail.com> Message-ID: <1669806522.2278161244166816193.JavaMail.root@zm06.stanford.edu> Hi, pquest users, There was an application error when I was starting a recipe. After I clear the error by terminating the application and rerunning it, the pressure became abnormal. The pressure couldn't be regulated to the setpoint but became higher and higher over time. I met this same problem ago and Jim Kruger suggested it might have been a pump being turned off when I terminated the application. I didn't know how to restart the pump so had to quit my run. Does anyone know how to make the system return to normal? I have reservation tomorrow morning and really hope somebody can fix the system by that time. Thanks! Ying From jimkruger at yahoo.com Fri Jun 5 19:22:15 2009 From: jimkruger at yahoo.com (jim kruger) Date: Fri, 5 Jun 2009 19:22:15 -0700 (PDT) Subject: Pquest Message-ID: <453018.62695.qm@web38901.mail.mud.yahoo.com> PQuest seems to be working OK except: still no chuck heaters but cooling flow seems better, i.e "plasma heating" seems less than before. RF power must be divided by about a factor of 2 again to get the same bias voltage as earlier this year, and what seems to be the same etch selectivity. This is a second factor of 2: a recipe that used 40 watts RF last year required only 20 watts after the Holiday shutdown modifications. After the latest repairs now the required RF is now 9 watts. He blow-by is bad on standard wafers (6 or 7, with 8 = max.). I found today that thick (850um), double polished wafers give a perfect seal. The He offsets have been zeroed. Perfect seal used to show 3 sccm He (so 11 sccm flow with no wafer). Now perfect is 0 sccm He and "no wafer" = 8. jim --- On Fri, 6/5/09, Andrei Faraon wrote: > From: Andrei Faraon > Subject: Pquest > To: jimkruger at yahoo.com > Date: Friday, June 5, 2009, 11:09 AM > Hi Jim, > > I just wanted to ask you about the current status for > pquest. Is it > operating OK? Any advice on the etching parameters? > > -- > Andrei Faraon > > Stanford University, Applied Physics > 316 Via Pueblo Mall > Stanford, CA, 94305 > Mobile: 650 714 7881 > Office: 650 723 2279 > www.stanford.edu/~faraon > From dlee76 at stanford.edu Tue Jun 9 13:59:38 2009 From: dlee76 at stanford.edu (Donkoun Lee) Date: Tue, 09 Jun 2009 13:59:38 -0700 Subject: Etching condition for Cr and SiO2 Message-ID: Hi all: I plan to seperately etch 100 nm-thick SiO2 and ~50 nm thick Cr layers. Does anyone have some experiences of the etching process for Cr and SiO2 layer? If you know, can I share your experience, such as ECR power, RF power, gas, flow rate, working pressure, and etching rate of each material? I really appreciate for your time and kind consideration, in advance. Kind Regards, Donkoun Lee -- Donkoun Lee Ph.D Candidate Prof. Shan X. Wang's Group Dept. of Materials Science & Engineering, Stanford University McCullough Building, Rm 208, 476 Lomita Mall, Stanford, CA, 94305-4045 Office: (650)-723-2939 Fax: (650)-736-1984 From shott at stanford.edu Fri Jun 12 17:34:02 2009 From: shott at stanford.edu (John Shott) Date: Fri, 12 Jun 2009 17:34:02 -0700 Subject: Weekend Update!!! Message-ID: <4A32F3FA.7080503@stanford.edu> Pquest Community: I believe that we have good news to report: as of 5:20 pm Friday evening, the pquest is under vacuum and we believe that we have addressed 3 significant issues: 1. We have replaced the ferrofluidic feedthrough that Jim Kruger reported to be leaking earlier this week. Fortunately, we had one of the spare, chemical-resistant ones on hand. 2. Elmer and Jim Haydon determined that the old belt driven assembly that drives the wafer clamp up and down was likely placing lateral torque on the ferrofluidic feedthrough .... resulting in shorter than expected life. They have done a nice (and quick) job of converting it to an on-axis direct drive system. Note: for those of you used to the sound of the "whine" as the clamp mechanism drives up and down, be forewarned: it's pretty silent now so you won't get the same audible cues. 3. In the process of addressing these issues during final testing, we experienced problems with the drive mechanism that moves the wafer arm to/from the load lock onto the chuck. We have disassembled and repaired the drive mechanism. Note: for those of you used to the sound of the "click, click, click" as the wafer arm pulls out of the etch chamber, it is now also pretty silent. No more audible cues there either. Cesar is monitoring the vacuum system now and will make a final check of wafer load/unload and wafer clamp operations and will also likely run an oxygen plasma as well. Barring any surprises, I fully expect that Cesar will release the tool for use by about 7 p.m. this evening. We are hopeful that this will allow some of you to get some useful etching done over the weekend. At this point, I believe that the only thing yet to be resolved is the delivery and installation of the chuck heaters .... so we do not yet have the electrically heated chuck. Other than that, however, I believe that the system to be fully functional. Good luck and happy processing, John From mihuhou at stanford.edu Sat Jun 13 11:09:40 2009 From: mihuhou at stanford.edu (Ying Chen) Date: Sat, 13 Jun 2009 11:09:40 -0700 (PDT) Subject: Pquest RFOUT In-Reply-To: <4A32F3FA.7080503@stanford.edu> Message-ID: <490628602.3791161244916580823.JavaMail.root@zm06.stanford.edu> Hi, PQ users, I was trying to use OXYGEN3.RCP to clean the chamber. But the RF was showing "RFOUT" with RF power reading <5 and the matching system always went out of limit. Does anyone know how to solve this? Thanks! Ying From mihuhou at stanford.edu Wed Jun 17 20:00:20 2009 From: mihuhou at stanford.edu (Ying Chen) Date: Wed, 17 Jun 2009 20:00:20 -0700 (PDT) Subject: Again. Re: Pquest RFOUT In-Reply-To: <490628602.3791161244916580823.JavaMail.root@zm06.stanford.edu> Message-ID: <738726643.4343791245294020228.JavaMail.root@zm06.stanford.edu> RFOUT again! Users before me, Was RF ok when you were using the pquest today? Thanks! Ying ----- Original Message ----- From: Ying Chen To: pquest at snf.stanford.edu Sent: Sat, 13 Jun 2009 11:09:40 -0700 (PDT) Subject: Pquest RFOUT Hi, PQ users, I was trying to use OXYGEN3.RCP to clean the chamber. But the RF was showing "RFOUT" with RF power reading <5 and the matching system always went out of limit. Does anyone know how to solve this? Thanks! Ying From senyod at yahoo.com Thu Jun 18 05:54:05 2009 From: senyod at yahoo.com (Senyo Dogbe) Date: Thu, 18 Jun 2009 05:54:05 -0700 (PDT) Subject: Again. Re: Pquest RFOUT Message-ID: <530639.99614.qm@web36106.mail.mud.yahoo.com> Ying, I used it to etch 3 wafers successfully. There was no RFOUT problem. Jim Krugger also did not have this problem. Senyo --- On Wed, 6/17/09, Ying Chen wrote: > From: Ying Chen > Subject: Again. Re: Pquest RFOUT > To: "Ying Chen" > Cc: pquest at snf.stanford.edu > Date: Wednesday, June 17, 2009, 8:00 PM > RFOUT again! > > Users before me, > Was RF ok when you were using the pquest today? > > Thanks! > Ying > > ----- Original Message ----- > From: Ying Chen > To: pquest at snf.stanford.edu > Sent: Sat, 13 Jun 2009 11:09:40 -0700 (PDT) > Subject: Pquest RFOUT > > Hi, PQ users, > > I was trying to use OXYGEN3.RCP to clean the chamber. But > the RF was showing "RFOUT" with RF power reading <5 and > the matching system always went out of limit. > > Does anyone know how to solve this? > > Thanks! > Ying > >