Shutdown raith SNF 2006-12-08 14:51:57: Electron optics board may have been damaged im EHT trip event.
jwc at snf.stanford.edu
jwc at snf.stanford.edu
Fri Dec 8 14:51:58 PST 2006
Good Afternoon:
Electron optics board in the RAITH 150 may have been damaged im EHT trip event that occurred early this morning. This trip may have been a result of touching the pole piece cap on the final objective lens during the sample holder crash. They occurred within a minute of each other on the LEOLOG.LOG.
I attempted to recalibrate the E/O board mag. amplifier and stigmation offset correction functions but I cannot achieve the larger range of control I need to be able to stigmate well enough to obtain a clear image and perform the calibration. The lens or stigmation quadropoles may have been damaged.
Spoke with Zeiss Field service by phone and we may be able to get a service person on site Monday afternoon next week.
Until then we are down and off line on the system. No user access to the RAITH or the LEO systems or PC's over the weekend.