Comment raith SNF 2011-08-26 15:09:00: we are now using recom until further notice. Use this procedure...

jwc at snf.stanford.edu jwc at snf.stanford.edu
Fri Aug 26 15:09:01 PDT 2011


We have worked out a alternative to resolve the API interface errors in the communications between computers that we have been experiencing this week.
We have now turned off the TCP/IP communications and resu,ed using REMCON.exe serial interface between the RAITH and LEO computers.
Here what to follow for success:
1. Start on Right hand computer we call LEO. Click desktop ICON called SMARTSEM or you can  press [START] + [PROGRAMS]+[SMARTSEM.exe] login as normal if you have established a new LEO account otherwise:
New Users use the training login passed newbie.
Experienced Users use the testuser login passwd test.
2. LEO computer [START] + [PROGRAMS]+[SMARTSEM]+[REMCON.EXE] to start the remcom serial comms. interface...
3. RAITH computer Press the RAITH150.exe 
Or [START] + [PROGRAMS]+[RAITH]+[RAITH150.EXE]
3. PROCEED AS NORMAL BUT MOST FUNCTIONS AND CONTROL CAN BE PERFORMED FROM THE RAITH SIDE.
MANUAL WRITEFIELD ALIGNMENT NOTES:
START WITH THESE ZOOM FACTORS:
0.24 U, 0.24 V FOR 10KV 10UM APERTURE @ 5 mm.
BE SURE TO UPDATE YOUR W BEFORE WRITE FIELD ALIGNMENT.
NEED HELP HIGHLIGHT THE MODULE YOU HAVE A QUESTION FOR AND PRESS [F1 KEY] Context sensitive help will come up and it searchable too!
Need more help or support call 631-738-9500 8 - 5 eastern Time for support and tell then you are calling from the ebeam lab... Speak with Cole Loomis, Joe Klingfus or Kevin Burcham with RAITH USA field service.