Raith: please don't use 15-20.9KV or 23-30KV
James Conway
jwc at snf.stanford.edu
Thu Apr 22 08:48:19 PDT 2004
Ji-Won is correct that the dangerous range is the ~18 - 20 keV. The
threshold for the arcing events we observed before was in the 26.5 keV
total acceleration voltage range. Keep in mind that the beam booster
section of the column accelerated electrons from the 8 keV primary
energy to the specified kEV through the 20 kEV range. At 20001 - 20100
Volts this section shuts off and ramps back to 0 and then without the
beam booster section applied accelerates electrons to the specified
higher acceleration voltage. So 20 kEV results in 28 kEV potential
between the gun tip and the anode plate.
I have requested LEO to return to assist in evaluating this problem, but
with no response thus far. I escalate this with LEO in Germany in the
morning Friday their time.
Mike Santomango with LEO has been MIA since I returned from Germany two
weeks ago, and I am continuing attempts to reach him, or someone with
suitable knowledge here in the USA.
Thank you for your support!
James Conway
Ji-Won Son wrote:
>Hi Mark,
>
>Thank you for the message. But one thing which is not clear to me is that
>why the range between 20.1-20.9 is included in the dangerous range. As far
>as I know, the most dangerous range is 15 to 20 kV, and 23 kV and higher
>(where the actual acc. voltage is higher than 23 kV). Does anyone experience
>problem in the range of 20.1 to 20.9kV?
>
>I would appreciate your help. Thank you.
>
>=======================
>Sincerely,
>Ji-Won Son
>
>Office: (650) 723-4940
>Fax: (650) 725-3459
>=======================
>
>
>-----Original Message-----
>From: Mark Topinka [mailto:mtopinka at stanford.edu]
>Sent: Wednesday, April 21, 2004 8:01 PM
>To: raith at snf.stanford.edu
>Subject: Raith: please don't use 15-20.9KV or 23-30KV
>
>(in other words, 0-15KV is ok and 21-23KV are probably ok, but DO NOT set
>it to 15-20.9KV or 23-30KV : these ranges cause intermittent EHT trips,
>which can occasionally turn into the more serious and dreading "arc", which
>can ultimately seriously damage the instrument) Sylvia had problems with
>repeated EHT trips (even after a LONG pumpdown) today and we do not want a
>repeat of what happened earlier this year (right!?) until we sort out what
>exactly is going on. -Mark
>
>
More information about the raith