Shutdown Raith SNF 2004-08-12 13:54:23: LEO E/O BOARD FAILURE...
James Conway
jwc at snf.stanford.edu
Thu Aug 12 14:21:08 PDT 2004
FWD'd from Raith-PCS.
In addition to the E/O Board replacement mentioned below, it appears
that the column has become contaminated recently.
This is evidenced by the increasing number of reports and severity of
the image drift after blanking and unblanking the beam which resulted in
most of the recent difficulties that Users have reported when performing
the align writefield routine.
Once we get ZEISS back with a E/O Board we can continue to troubleshoot
the system and effect further maintenance and repairs as necessary.
Further updates as we have information.
Thank you,
James Conway
-------- Original Message --------
Subject: Shutdown raith SNF 2004-08-12 13:54:23: LEO E/O BOARD FAILURE
Date: Thu, 12 Aug 2004 13:54:24 -0700
From: jwc at snf.stanford.edu
To: raith-pcs at snf.stanford.edu
The scan rotation calibration function used to maintain Orthogonality is failing -- slowly, and now has failed.
This is the source of the misshaped shot spots in the V (Y) axis imaging plane.
This will require replacement of the E/O Board.
We are attempting to locate a calibrated and tested board from ZEISS.
James Conway 08122004_1351 hours.