Double layer PMMA dose

James Conway jwc at snf.stanford.edu
Thu Jan 29 16:45:26 PST 2004


Hi Scott:

Hitachi 30 KeV:

Threshold to clear large areas:  350 uC/cm2
Normal area dose I use:  500 uC/cm2

Line dose to 100 - 150 nm: 650 - 800 uC/cm2   Jing Kong used 900 uC/cm2

You will have to adjust your dose for 10 keV downward divide by 2.5.
10 keV gave me ~ 0.2 um of undercut.

I will be running more samples in late February.

All the best,

JWC




Scott D. Andrews wrote:

>Hello,
>
>Does anyone have recent data on the proper dot and line dose for the
>bilayer PMMA process?  If you also know the undercut you achieved, could
>you let me know that as well?
>
>Thank you,
>Scott Andrews
>  
>