RAITH status Monday March 1, 2004: 17:21 hours.

James Conway jwc at snf.stanford.edu
Mon Mar 1 17:32:54 PST 2004


Greetings,

Neither Mike Santomango nor I were pleased with the emission and beam 
profiles of the FE-GUN Tip as we arrived in this morning.  While the 
emission is near normal levels we had to drive the FE-gun far over 
specifications to achieve the normal level. EHT V at 6400 V instead of 
the normal 4400 - 5400 V.

We have decided the best approach at this point is to simply replace the 
FE-GUN tip with a new clean unused tip and perform another bake out and 
run up sequence.  While we were open we also cleaned the Extractor plate 
of Hydrocarbon residue that appeared over the weekend.

  We just started pumping and will perform an extended bake out until 
WEDNESDAY morning.  Wednesday afternoon we will perform alignments and 
adjustments and will again attempt to qualify the systems for line width 
and stitching performance.  I would expect access to reopen to users 
Thursday morning.  Please adjust your plans accordingly.

Thank you,

James Conway