Important: High voltage clarification
Ji-Won Son
jwson at stanford.edu
Thu Mar 18 10:27:30 PST 2004
Hi all,
I asked the procedure to use voltage around 20keV, so I think I can answer
some of David's question.
The Raith system use 8kV deceleration up to 20keV, so the actual voltage
that applied to the gun is plus 8V to the acc. voltage you choose. (if you
use 20keV, then the actual voltage is 28 V, not 20.)
Above 20kV, the system turns off 8V deceleration part and applied voltage
will be same as you choose. That's why James asked us to use 10-15 kV and
20.1 - 21 kV at this point to reduce the load.
I am not sure about using 10-15kV, but when you use 20.1-21kV, the procedure
is:
1. Turn the beam on at 10kV.
2. Check the vacuum. (should be around 10^-9)
3. If vacuum is ok, then change the voltage to 20.1-21kV, by one step. Don't
ramp up slowly step by step!
These are what I heard from Raith field service. James, if there is anything
wrong with these, please correct.
Thank you.
====================
Sincerely,
Ji-Won Son
(Office) 650-723-4940
(Fax) 650-725-3459
====================
-----Original Message-----
From: David Goldhaber-Gordon [mailto:goldhaber-gordon at stanford.edu]
Sent: Thursday, March 18, 2004 6:56 AM
To: 'James Conway'; 'John Cumings'
Cc: 'Raith SNF Mailing list'
Subject: Important: High voltage clarification
Hi all,
Thanks and congratulations are due to James for pushing so hard to get the
system up before he leaves for Germany.
I want to clarify one point: James probably meant that the voltages me may
use are 10-15 kV or 20.1 to 21 kV (i.e. 20100 to 21000 V). I point this out
since the actual numbers may be important and I'd hate for someone to
misunderstand and jeopardize the system for him- or herself and others.
James, please post a correction if I got this wrong. Also, could someone
comment on the proper procedure for getting up to those voltages? I'm not
sure that procedure has been uniformly agreed upon and communicated.
Best,
David
----------------------------------------------------------------------------
David Goldhaber-Gordon goldhaber-gordon at stanford.edu
Assistant Professor of Physics davidg at post.harvard.edu
Stanford University (permanent forwarding)
www.goldhaber-gordon.com
(650) 725-2047 (lab) (650) 724-3709 (office)
Address for letters or packages: Administrative Associate:
David Goldhaber-Gordon Roberta Edwards
Geballe Laboratory for Advanced Materials McCullough, Rm. 338
McCullough Building, Room 346 Phone: (650) 723-8028
476 Lomita Mall Fax: (650) 724-3681
Stanford, CA 94305-4045 email: redward at stanford.edu
-----Original Message-----
From: James Conway [mailto:jwc at snf.stanford.edu]
Sent: Wednesday, March 17, 2004 8:25 PM
To: John Cumings
Cc: Raith SNF Mailing list
Subject: Re: Raith is being qualified... Thread: Required Limits of keV
operation through 6APRIL2004 on RAITH 150
Please Everyone:
We would like to operate the system in a range between 10 - 15 KeV and/or
at 20100 to 2100 kV only. NOT 20 kV or higher than 21 kV.
When I return from Germany and with LEO field service we will perform
further testing on the column.
Things are in good order and tomorrow I plan to put the system back into
green light mode as more measurements and imaging is completed
Stitching Mean + 3 sigma was 19 nm.
Thanks to everyone for your support during this challenging period,
JWC
John Cumings wrote:
James,
Green light means 30 kV is okay.
I presume you mean yellow light, because 30 kV is not okay, correct?
-John
------------------------------------------------------------
John Cumings
cumings at stanford.edu
Postdoctoral Scholar
Physics Department
Stanford University
office (650) 725-1025
lab (650) 725-2047
fax (650) 724-3681
Geballe Laboratory for Advanced Materials
239 McCullough Building
476 Lomita Mall
Stanford, CA 94305-4045
------------------------------------------------------------
In message <4058B6CE.7060400 at snf.stanford.edu>you write:
<!DOCTYPE html PUBLIC "-//W3C//DTD HTML 4.01 Transitional//EN">
<html>
<head>
<meta http-equiv="Content-Type" content="text/html;charset=ISO-8859-1">
<title></title>
</head>
<body text="#000000" bgcolor="#ffffff">
<br>
Greetings:<br>
<br>
I am most relieved to report that we are coming back into normal
operational status on the RAITH 150 system.<br>
<br>
This afternoon and evening we will be qualifying the system and working
with users whom are in crisis mode.<br>
<br>
We hope to go into green light mode in the next twelve hours.<br>
<br>
James Conway<br>
</body>
</html>