Will be Raith be up on Friday morning? YES! Current PLAN -- system coming back on line 1800 Thursday
James Conway
jwc at snf.stanford.edu
Wed May 5 17:14:15 PDT 2004
Greetings:
Carl Zeiss and RAITH Field Service have just handed the system back to
me for performance testing and LEO column adjustments evaluation.
My plan:
1. Complete beam drift evaluations this afternoon. Beam drift is
well within specification (spec. 10 nm/min. measured 3 - 5 nm/min.
2. Hand off the system to Gigi and other available RAITH champions
after GiGi overnight and into Thursday morning.
3. I will return to the system at 10 AM tomorrow and hope to complete
qualification testing by late tomorrow afternoon.
4. I hope to release the system back to users at 6 PM tomorrow.
Thank you for your patience and support!
James Conway
Peter Catrysse wrote:
>Hi James:
>
>I am losing two 4-hour time slots next week (I had Tuesday and Thursday
>morning) as a result of the training class you are giving. I reserved some
>time this Friday morning. How realistic is it that you will be done by then?
>
>Thanks,
>
>Peter Catrysse
>
>
>
>>-----Original Message-----
>>From: James Conway [mailto:jwc at snf.stanford.edu]
>>Sent: Wednesday, May 05, 2004 1:21 PM
>>To: Peter Catrysse; Raith SNF Mailing list; ebeam at snf.stanford.edu
>>Subject: I have a 6 inch wafer HEADWAY chuck you can check out from me for
>>EBeam Lab use.
>>
>>Greetings:
>>
>>I have now obtained a 6 inch wafer chuck for the Headway spin coater.
>>This is to be kept secured at my desk and will need to be check out
>>through me for every loan.
>>
>>Thank you for your support!
>>
>>James Conway
>>
>>
>>
>
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20040505/70b04c87/attachment.html>
More information about the raith
mailing list