Raith Status Friday Septrember 17, 2004 1330 hours.
James Conway
jwc at snf.stanford.edu
Fri Sep 17 14:15:13 PDT 2004
Greetings Raith Community:
I had a large amount of time on the system this week and was able to
complete two difficult pattern writes over the last two days. These are
the wafers that I had been trying to write for nearly a month. A number
of simple test and check outs were performed in parallel to this work.
The RAITH 150 system is working well and appears to be within
specification for stitching and better than specification for overlay. I
will be repeating the PCM performance testing we completed August 28,
2004 to confirm we are still were we were then. As this work is
completed I will report the results and post images on the bulletin
board superseding the work done in May 2004.
LEO /Zeiss SMT will return early next week for adjustments to and
checking the settings on the E/O and column calibrations we performed
two weeks ago. We will also attempt to bring down the Zoom V value to
levels farther from the Zoom V Limit value. I have been able to complete
a number (9X3) write field alignments without encountering the Sweep Sum
exceeds limits error. Several users whom encountered problems during
write field alignment had not measured and adjusted the WD to 5 mm using
the SEM to measure WD. IN one case they were only at 3 mm!
The FE-Gun tip has run in nicely and Gun vacuum EHT ON is in the 8E-010
T region.
The system has been tested this week to 25 kV with extended writes
performed at 20 and 10 keV. I will perform some pattern writes next
week at 20,25, and 30 kV to ensure beyond any doubt that we can start
general operations at this acceleration voltage for any users wishing to
do so.
I recalibrated the Piezo height controls using a bare Si wafer on the
ElectroStatic Chuck (ESC) and the new calibration for centered post is
within 0.149 mm across a six inch wafer. I will be making mechanical
adjustments to the ESC and will recalibrate the piezo's once again. At
that point we should be able to level samples on the ESC and not exceed
the range of the piezo's travel in Z. Following that I will evaluate
the leveling of the Standard Sample holder as it sits onto the kinetic
mounts of the Piezo post and see if there are actually any issues with
this holder being level or not.
Made measurements of the Beam Current at 10 keV and 5 mm Working
distance for all the apertures.
We are in the middle of the decent of Beam Current as we close these
apertures with hydrocarbon contamination. Overall we are doing well on
this issue and beam speeds are near optimum for all apertures on 100 nm
thick 2% 950K MW PMMA; allowing us to use suitably small area steps and
still have Beam speed below 10 mm/sec.
Beam Current (Beam I) measurements I made today:
APT Beam I Today Beam I after May PM and aperture
plate replacement Delta
(um) ( nA ) (nA aperture values
just installed) (% of Beam I new)
120 2.345 4.842
48
60 0.6182 1.025
60
30 0.1339 0.233
57
20 0.0699 0.085 6-15-2004
first value recorded in logbook? 82
10 0.0168 0.0459
36
7.5 0.0078
0.011
70
Conclusions: We are approximately 50 % into our PM interval with
acceptable beam currents for all apertures albeit slightly on the low
side for the 10 um aperture. This is consistent with the usage we have
on the system for each aperture, particularly considering the large
number of EBL writes that have been performed this summer.
The 10 um aperture beam I is reducing more rapidly than during the
previous PM interval last year. This value may be indicative of a
greater contribution of contamination from samples by users working with
this aperture than the 30 um aperture.
Thanks to the many users which have been diligently recording their Beam
Currents in the logbook. This information is very useful to us and
helps me greatly to monitor the condition of the system.
System is up and in Green Light for your usage. If anyone is having any
problems whatsoever on the system please see me during my office hour.
Thank you for your support and have a pleasant weekend.
James Conway
Upcoming in October:
RAITH Group XI a four day intensive short course will be conducted the
week of October 12 - 15, 2004
There are two slots remaining in this class. Please maximize your
utilization of the system coming into this period.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20040917/9f715615/attachment.html>
More information about the raith
mailing list