Raith: interesting (important?) discovery
James Conway
jwc at snf.stanford.edu
Tue Jul 5 12:15:51 PDT 2005
Hello mark and RAITH Users:
This is the first report of this kind of problem. Has anyone else seen
this type of problem?
Please check your time delay settings to minimize dynamic effects:
In the exposure control box check DETAILS. This is the time delay
between stage moves and Write field to Write Field stitching moves
between write fields. Best value is 5 ms (default), but it is user
adjustable and it may have been changed in the specific users' login.
Times greater than 3 ms have been found to expose correctly.
Thank you,
JWC
Mark Topinka wrote:
> Hi- if anyone is continuing to have problems with part(s) of their
> pattern being blurry or not exposing at all, while the rest of it
> seems fine, please contact me asap. I think I've figured out a
> work-around, even though I still don't understand the underlying
> hardware problem. As best I can tell, this is not happening because
> of a leaky beam blanker. Rather, it seems that after a long time
> being blanked (like, between writefield moves), the beam is out of
> focus and/or low in beam current for the first roughly 0.01-0.1
> seconds after being unblanked (+ or - a factor of ten, probably - I
> haven't measured it carefully, and it probably depends on lots of
> things). Short blankings, like between dots or objects within one
> writefield don't affect it, so after the initial "warm-up" time in
> each writefileld, my experience has been that the rest of the pattern
> writes great. The (hack) solution that I found is to make a
> sacrificial object in each writefield that takes about a second to
> write and is the first thing that gets written. (Use the "O" command
> in the raith pattern editor to order your objects so that it is the
> first object written, otherwise it does no good). I have no idea
> *why* the raith is doing this. Has anybody else continued to
> experience these problems (even after the column replacement?) or have
> I been the only one? -Mark
>
More information about the raith