NEW LEO SEM COLUMN installed and system qualified, now back on-line. Users restricted to 10 kv through July 6, 2005

James Conway jwc at snf.stanford.edu
Fri Jun 24 18:31:54 PDT 2005


      Greetings Raith Community:


This week we received installed and characterized a brand new LEO 1530 
SEM column on the RAITH 150 system.
Tuesday evening, working into the early morning with Jason Sanabia and 
Mirwais Aktary, we exchanged the SEM column on the Raith 150 system with 
a new SEM column. Through the remainder of the week we have been 
performing many test and qualification evaluation runs.

I am very much relieved to report that the blanker leakage and column 
scattering issues that we have experienced since mid May of last year 
have now been fully resolved to my satisfaction.  Emission and scatter 
from the column when blanked on the Raith Blanker are now between 0.0002 
and 0.0003 nA for all apertures.  There is no difference between this 
value when the LEO blanker is enabled either. A test script that we have 
been employing to evaluate scatter for various impingement periods, 
which well displayed the problem before we installed the new column, 
also displayed no exposure artifacts evident whatsoever after the column 
exchange.

Please READ:
WE NOW HAVE BASICALLY A BRAND NEW AND CALIBRATED RAITH 150 SYSTEM.  THE 
SEM COLUMN, IN-LENS DETECTOR, THE FE-GUN, AND THE BLANKER SECTIONS ARE 
BRAND SPANKING NEW! THE SYSTEM AND COLUMN ARE EXQUISITELY CLEAN AND FREE 
OF CONTAMINATION -- WE MUST ALL STRIVE TO DILIGENTLY KEEP IT THIS WAY.

ALL USERS MUST COMPLY WITH THE FOLLOWING PROTOCOLS EXACTLY WITH NO 
EXCEPTIONS:

   1. NO OPERATIONS ABOVE 10 kV ACCELERATION VOLTAGE WHILE WE CONDITION
      THE FE-GUN.( through 7-6-1005)
   2. NO RESIST OR OTHER MATERIALS THAN HAVE NOT BEEN SOFT BAKED TO >
      150 DEGREES C.
   3. BEFORE GOING INTO THE LOAD LOCK ON THE SYSTEM ALL CHIPS AND WAFERS
      MUST BE SOFT BAKED AGAIN AT >= 90 DEGREES CELSIUS FOR 90 SECONDS.
      REMARKABLE YOU WILL OBTAIN BETTER CONTAMINATION DOTS TOO!
   4. NO CARBON TAPE, EPOXIES, GREASES, OR OTHER ADHESIVES ARE TO BE
      PLACED INTO THE SYSTEM OTHER THAN MINUTE FRAGMENTS OF SPECTRA TABS
      THIN CONTAMINATION FREE CARBON DOTS. SMALL 1 MM STRIPES ARE
      RECOMMENDED.  ALL OTHER HOLD DOWN FIXTURES MUST BE FRESHLY SOLVENT
      CLEANED AND BAKED BEFORE ENTERING SYSTEM.

IF YOU ARE UNABLE TO FOLLOW THESE GUIDELINES EXACTLY TAKE YOUR WORK 
ELSEWHERE!

Should you have any questions please see during my office hour!


      Finally everyone working on the Stanford RAITH system owes a great
      debt of gratitude to the following people for making the extra
      effort and going the distance to resolve this problem.  Any other
      manufacturer would have not gone to this level of effort that the
      RAITH team has made to address our concerns, and to ensure that we
      had the very best system possible to perform our research here at
      SNF! 


        Special recognition goes out to:

George Lanzarotta gl at raithusa.com President RAITH USA  -- This 
replacement is likely costing him in excess of $50 K US.

Ralf Jede Director RAITH GmbH for his continuing concern and support 
through this period. He freed up the personnel resources necessary to 
successfully complete the task from the RAITH Dortmund site side.

Bernd Stegemann, Stegemann at Raith.de   RAITH GmbH, Dortmund.  For time 
and travel well above reasonable expectations. His outstanding talent 
and effort made both here at Stanford and at Raith in Dortmund, DE and 
working in Oberkocken at the Zeiss factory ensured we obtained the very 
best column set up possible.

 Michael  Kahl kahl at raith.de for his continuing technical and 
application support.  We worked well into the night during my visit to 
Dortmund, DE in April and we made critical observations comparing 
several RAITH systems that gave us the right directions to proceed IN to 
address the scattering issues.

Cole Loomis RAITH USA Field Service Manager  We must have had  100 hours 
on the phone CALLS nights and mornings during measurements over the last 
year.

Jason Sanabia and Mirwais Aktary RAITH USA Field Service for time and 
effort above and beyond any reasonable expectation.  The three of us 
have been at it since the minute they arrived here at SNF this week and 
we completed the plan on time and with an successful outcome. These guys 
are "The Right Stuff."

Mike Santomango Zeiss SMT Field Service.  Despite other commitments from 
Zeiss, he did his absolute best and made sure that when we needed him he 
was there and was able to get the adjustments performed that needed to 
be performed, and for his continuing support. He is one of Zeiss's very 
best technical people.

Dan Pickard, Pease Group Stanford. For continuing consultation and 
support related to electro-static and SEM Column Theory in discussions 
that were useful and informative toward reaching a solution. 


There were many other people involved in this work over the last year 
from both RAITH organizations, Zeiss SMT, and Stanford. Please accept my 
heartfelt thanks for your encouragement and support through the past 
year.  Many many thanks to you all.

Now lets get back to our research and together let us push the limits of 
Ebeam Lithography here at Stanford Nanofabrication Facility.

Sincerely Yours,

James Conway
Ebeam Technology Group
Stanford Nanofabrication Facility








-------- Original Message --------
Subject: 	Problem raith SNF 2005-06-24 13:36:06: Users restricted to 10 
kv through July 6, 2005
Date: 	Fri, 24 Jun 2005 13:36:07 -0700
From: 	jwc at snf.stanford.edu
To: 	raith-pcs at snf.stanford.edu



No operations above 10 kV -- the system was tested hot to 20 kV and cold to 30 kV but we wish to allow additional time for FE-Gun tip conditions to stabilize and vacuum to enter the 10e--010 mbar range.  
I will qualify the system for higher acceleration voltages on July 5, 2005
James Conway
06242005:13:34:00.00