Kenneth Chen kychen51 at yahoo.com
Sat Mar 5 08:55:12 PST 2005


The random jumping is still occurring after the system has been enabled over a hour.  

With the standard configuration of 10kV, 30 aperature, 550x mag, beam current 0.1803nA,  the random jumping occurs between 13.35kX - 13.5kX.  If you manually adjust the column control to skip past that range, to say...20kX, sometimes the shifting still occurs, but sometimes there's no shifting.  The shift is approximately one viewing area to the lower left.
 
This shifting causes problems with the align writefield as you don't end up in the spot you were previously focusing and stigmating on.  I also cancel the write as I can't get past this step.
 
- Kenneth Chen

Ilya Fushman <ifushman at stanford.edu> wrote:
There was an EHT trip when I tried to go to 20kV from 15kV. As a result, the
settings on the in-lens detector are off and when I try to go past ~ 15kX on
the magnification, the image area jumps to some random part. At this point
it's impossible for me to stig and focus. According to James the system
should be OK after 1 hr or so. I have another commitment and cannot stay
longer to find out.

-Ilya

---------------------