From jwc at snf.stanford.edu Fri Jul 7 17:20:23 2006 From: jwc at snf.stanford.edu (James Conway) Date: Fri, 07 Jul 2006 17:20:23 -0700 Subject: [Fwd: Comment raith SNF 2006-07-07 17:10:04: Raith is qualified] Message-ID: <44AEFA47.50108@snf.stanford.edu> Greetings Raith Users: System has passed qualification with flying colors. Excellent spot and SPLine resolution with short shots to 13 nm. Measurements on this test next week. Users should not operate above TEN KV acceleration voltage. Thank you for your support! James Conway P.S., on the road this weekend available on my cell phone... -------- Original Message -------- Subject: Comment raith SNF 2006-07-07 17:10:04: Raith is qualified Date: Fri, 7 Jul 2006 17:10:05 -0700 From: jwc at snf.stanford.edu To: raith-pcs at snf.stanford.edu qual test and pcm test run with excellent results. System back up online and available for users as scheduled. Users restricted to no more than 10 kV Acceleration voltage for the next week as we run in the NEW FE-GUN. James -------------- next part -------------- An HTML attachment was scrubbed... URL: From kimsangb at stanford.edu Sat Jul 8 03:13:08 2006 From: kimsangb at stanford.edu (SangBum Kim) Date: Sat, 8 Jul 2006 03:13:08 -0700 Subject: raith free until 10:30AM. Message-ID: <003c01c6a277$1c40e220$a9b50c80@anavel> Could not get the sample ready. Sorry for late notice. -------------- next part -------------- An HTML attachment was scrubbed... URL: From mtang at stanford.edu Mon Jul 10 07:51:18 2006 From: mtang at stanford.edu (Mary Tang) Date: Mon, 10 Jul 2006 07:51:18 -0700 Subject: Cancelled Reservation 1-6 today Message-ID: <44B26966.3020307@stanford.edu> Hi everyone -- James just called in sick, so I cancelled his reservation for training today. (And if you happen to know who the training session was for, please let me know so I can email him/her as well.) Thanks -- Mary -- Mary X. Tang, Ph.D. Stanford Nanofabrication Facility CIS Room 136, Mail Code 4070 Stanford, CA 94305 (650)723-9980 mtang at stanford.edu http://snf.stanford.edu From jwc at snf.stanford.edu Tue Jul 11 12:17:40 2006 From: jwc at snf.stanford.edu (James Conway) Date: Tue, 11 Jul 2006 12:17:40 -0700 Subject: Raith is qualified: Question on higher acceration voltage operations. Users restricted to below 20 kV until further notice. In-Reply-To: <6.1.2.0.2.20060708234856.05b1dc90@snowboard.stanford.edu> References: <44AEFA47.50108@snf.stanford.edu> <6.1.2.0.2.20060708234856.05b1dc90@snowboard.stanford.edu> Message-ID: <44B3F954.7020606@snf.stanford.edu> Ofer, Currently the system is not to be operated above 10 kV until I test the column at higher acceleration voltages. This week I hope to obtain stables gun vacuum levels and bring the system to 20 kV for testing. I will wish to maintain below 20 kV for at least the next ten days to two weeks. Once I know the FE-GUN is run in well , that emission is stable and that gun vacuum is fine, I will then test at elevated acceleration voltages and up to 30 kV. Please see me if you have any questions... Thank you, James Conway Ofer Levi wrote: > Hi James, > Thanks for the update. Glad to see that all goes well, with the Raith > maintenance. > As you recall, lately I use 30 KV in my writes. I reserved some time > for a week from today on the Raith. Can you please estimate if you > think by that time the Raith will be qualified for operation at 30 KV ? > > Thanks, > Ofer > > At 05:20 PM 7/7/2006, you wrote: > >> Greetings Raith Users: >> >> System has passed qualification with flying colors. Excellent spot >> and SPLine resolution with short shots to 13 nm. >> Measurements on this test next week. >> >> Users should not operate above TEN KV acceleration voltage. >> >> Thank you for your support! >> >> James Conway >> >> P.S., on the road this weekend available on my cell phone... >> >> >> -------- Original Message -------- >> Subject: Comment raith SNF 2006-07-07 17:10:04: Raith is qualified >> Date: Fri, 7 Jul 2006 17:10:05 -0700 >> From: jwc at snf.stanford.edu >> To: raith-pcs at snf.stanford.edu >> >> >> >>qual test and pcm test run with excellent results. >>System back up online and available for users as scheduled. >>Users restricted to no more than 10 kV Acceleration voltage for the next week as we run in the NEW FE-GUN. >>James >> >> > ______________________________________________ > > Ofer Levi, Ph.D. > Department of Electrical Engineering, Stanford University > CIS-X Rm 310, Stanford, CA 94305-4075 > > Phone: (650)723-0464 or 725-6907 > Fax: (650)723-4659 > Adm. Asst.: Gail Chun-Creech Ph: (650)723-0983 > E-Mail: levi at snow.stanford.edu > Web page: http://snow.stanford.edu/~levi/ > ______________________________________________ > -------------- next part -------------- An HTML attachment was scrubbed... URL: From cvenky at stanford.edu Tue Jul 11 18:33:41 2006 From: cvenky at stanford.edu (Venkatesh Chembrolu) Date: Tue, 11 Jul 2006 18:33:41 -0700 Subject: raith free tomorrow 4.30-7.00am Message-ID: <1152668021.44b4517590cd0@webmail.stanford.edu> Unexpected problems in sample preparation.... From jwc at snf.stanford.edu Wed Jul 12 17:29:32 2006 From: jwc at snf.stanford.edu (James Conway) Date: Wed, 12 Jul 2006 17:29:32 -0700 Subject: Will be finished earlier than I expected... Short NOTICE : Raith available after Randalls has finished his inspections about 7 - 8 PM. Message-ID: <44B593EC.1050308@snf.stanford.edu> Greetings: I am handing off the remainder of my raith session through Midnight tonight to User Randalls (Randall Stoltenberg). My write test just finished and if Randy doesn't want to use the system it is available -- otherwise available after Randalls is finished. I have more CAD work to complete before I can resume that particular projects writing and will resume these task tomorrow. Thank you for your support! James Conway From rmstolt at gmail.com Wed Jul 12 17:42:03 2006 From: rmstolt at gmail.com (Randy Stoltenberg) Date: Wed, 12 Jul 2006 17:42:03 -0700 Subject: Will be finished earlier than I expected... Short NOTICE : Raith available after Randalls has finished his inspections about 7 - 8 PM. In-Reply-To: <44B593EC.1050308@snf.stanford.edu> References: <44B593EC.1050308@snf.stanford.edu> Message-ID: <8769af0c0607121742k1c475e7dyba18f803af476f2e@mail.gmail.com> I will use the system starting 6:15 pm, but I should be done around 9pm. Randy On 7/12/06, James Conway wrote: > > > > Greetings: > > I am handing off the remainder of my raith session through Midnight > tonight to User Randalls (Randall Stoltenberg). > My write test just finished and if Randy doesn't want to use the system > it is available -- otherwise available after Randalls is finished. > I have more CAD work to complete before I can resume that particular > projects writing and will resume these task tomorrow. > > Thank you for your support! > > James Conway > > > -- Randy Stoltenberg Ph.D. Student, Department of Chemistry Stanford University Stauffer III, Room 15 381 North South Mall Stanford, CA 94305 randalls at stanford.edu Phone: 650 725 5403 Cell: 650 796 4527 Fax: 650-723-9780 -------------- next part -------------- An HTML attachment was scrubbed... URL: From jhemanth at stanford.edu Wed Jul 12 17:49:19 2006 From: jhemanth at stanford.edu (Hemanth Jagannathan) Date: Wed, 12 Jul 2006 17:49:19 -0700 Subject: Will be finished earlier than I expected... Short NOTICE : Raith available after Randalls has finished his inspections about 7 - 8 PM. In-Reply-To: <8769af0c0607121742k1c475e7dyba18f803af476f2e@mail.gmail.com> Message-ID: <000801c6a616$2cef2800$0100000a@jhemanthibm1> I will take the time from 9pm to midnight. Thanks Hemanth Jagannathan PhD Candidate Department of Electrical Engineering Stanford University, CA Phone :(O)(650)-725-0417; (C) (650)-380-2105 e-mail: jhemanth at stanford.edu -----Original Message----- From: Randy Stoltenberg [mailto:rmstolt at gmail.com] Sent: Wednesday, July 12, 2006 5:42 PM To: James Conway Cc: Raith SNF Mailing list Subject: Re: Will be finished earlier than I expected... Short NOTICE : Raith available after Randalls has finished his inspections about 7 - 8 PM. I will use the system starting 6:15 pm, but I should be done around 9pm. Randy On 7/12/06, James Conway wrote: Greetings: I am handing off the remainder of my raith session through Midnight tonight to User Randalls (Randall Stoltenberg). My write test just finished and if Randy doesn't want to use the system it is available -- otherwise available after Randalls is finished. I have more CAD work to complete before I can resume that particular projects writing and will resume these task tomorrow. Thank you for your support! James Conway -- Randy Stoltenberg Ph.D. Student, Department of Chemistry Stanford University Stauffer III, Room 15 381 North South Mall Stanford, CA 94305 randalls at stanford.edu Phone: 650 725 5403 Cell: 650 796 4527 Fax: 650-723-9780 -------------- next part -------------- An HTML attachment was scrubbed... URL: From ryantu at stanford.edu Thu Jul 13 08:25:15 2006 From: ryantu at stanford.edu (Ryan Tu) Date: Thu, 13 Jul 2006 08:25:15 -0700 Subject: Raith free Message-ID: Couldn't get samples ready in time this morning. From lxuwind at stanford.edu Thu Jul 13 21:26:08 2006 From: lxuwind at stanford.edu (Liang Xu) Date: Thu, 13 Jul 2006 21:26:08 -0700 Subject: Raith free 18:00-21:00 on Sat Message-ID: <002101c6a6fd$a131ccd0$38aa0c80@Leon> Liang Xu PhD candidate in Department of Materials Science and Engineering, Stanford University Tel: 650-723-2939 (office) 650-497-1934 (home) Address: 60 Olmsted Rd, Apt 104, Stanford, CA94305 -------------- next part -------------- An HTML attachment was scrubbed... URL: From jwc at snf.stanford.edu Fri Jul 14 15:37:25 2006 From: jwc at snf.stanford.edu (James Conway) Date: Fri, 14 Jul 2006 15:37:25 -0700 Subject: Users limited to no more than 15 kV on Raith. Message-ID: <44B81CA5.6020609@snf.stanford.edu> Greetings Raith Community: Today I tested the RAITH to 15 kV with no problems encountered while maintaining excellent Gun vacuum levels. (4.6 E-010 Torr. EHT ON at 15 kV.) All Users can now operate the system to acceleration voltages up to and including 15 kV for their writes. Higher acceleration voltages will be tested near the end of next week if possible. ++++ All Users are requested to pre-bake all their previously coated samples on a hot plate to at least 90 degrees C. before loading into RAITH ++++ Please take care to ensure that no oils, greases, waxes, or other contaminants remain on your samples when you load them into the system. Electron Beam diameter measurements at 10 kV 30 um aperture at 5 mm WD = 1.48 nm and at 15 kV, 20 um aperture and 4.93 mm WD is less than 1 nm. Patterns written during the qualification run last Friday displayed excellent stitching and overlay registration. 10 kV Dose to clear 100 nm 2% 950K MW PMMA in anisole films on Silicon was 110 uC/cm^2 for areas and 330 uC/cm for Single Pixel Lines. Single Pixels lines in the 15 - 25 nm range with good line edge roughness were obtained. Elevated vibration levels in the lab have been observed during hours of construction work on the project across the street. FWHM image ripple measurements on edge of 90 nm diameter latex spheres typically are 4 - 7 nm. During much of this week we have seen transient events to as much at 22 nm with a range of 7 - 15 nm typical. I will be reviewing the TMC / Raith gimbaled passive isolation platform tuning Monday afternoon to determine if I can reduce the impact of these vibrations. If you have any questions please see me during my daily office hour. Thank you for your support! James Conway SNF -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: blutextb.gif Type: image/gif Size: 569 bytes Desc: not available URL: From ycjun at stanford.edu Fri Jul 14 17:43:32 2006 From: ycjun at stanford.edu (Young Chul Jun) Date: Fri, 14 Jul 2006 17:43:32 -0700 Subject: Raith free Tomorrow (from Sat. 1pm) Message-ID: <1152924212.44b83a343d0c3@webmail.stanford.edu> Sample is not yet ready. I cancelled my Raith session tomorrow(Sat). - ycjun From jwc at snf.stanford.edu Mon Jul 17 11:47:10 2006 From: jwc at snf.stanford.edu (James Conway) Date: Mon, 17 Jul 2006 11:47:10 -0700 Subject: IMPORTANT: EHT Trip was encountered this weekend on RAITH at 15 kV. Please Bake your samples!!!! Message-ID: <44BBDB2E.5050102@snf.stanford.edu> An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: storm_Clouds.jpg Type: image/jpeg Size: 5175 bytes Desc: not available URL: From bcgierhart at ucdavis.edu Mon Jul 17 23:05:13 2006 From: bcgierhart at ucdavis.edu (Brian Gierhart) Date: Mon, 17 Jul 2006 23:05:13 -0700 (PDT) Subject: raith done early (11pm Mon) Message-ID: <200607180605.k6I65D7M011181@tremex.ucdavis.edu> Hi, I am done a little early with the Raith & it's free now. Regards, Brian From ycjun at stanford.edu Tue Jul 18 00:06:32 2006 From: ycjun at stanford.edu (Young Chul Jun) Date: Tue, 18 Jul 2006 00:06:32 -0700 Subject: raith done early (11pm Mon) In-Reply-To: <200607180605.k6I65D7M011181@tremex.ucdavis.edu> References: <200607180605.k6I65D7M011181@tremex.ucdavis.edu> Message-ID: <1153206392.44bc88786d8ef@webmail.stanford.edu> I'll take it. Thanks.. Quoting Brian Gierhart : > > Hi, I am done a little early with the Raith & it's free now. Regards, > Brian > From jwc at snf.stanford.edu Tue Jul 18 17:29:31 2006 From: jwc at snf.stanford.edu (James Conway) Date: Tue, 18 Jul 2006 17:29:31 -0700 Subject: Comment raith SNF 2006-07-18 16:55:08: System tested for operations at 5, 10, 15, 20, 25, and 30 kV] Message-ID: <44BD7CEB.3020903@snf.stanford.edu> Greetings raith Users: Today I tested the RAITH system at voltages ranging from 5 to 30 kV in 5 kV steps. No significant increases in system or FE-Gun section vacuum pressures where observed, other than the normal small increase and immediate decrease to base pressure when the Liner Voltage for the Beam Booster shuts off and ramps up above 20 kV. User restrictions on Acceleration Voltage are now increased to no higher than 20 kV. All Users should Pre-Bake all samples on a hot plate at or above 90 degrees Celsius just before you load into the system. Users wishing to work at higher acceleration voltages than 20 kV should see me so we can schedule a RAITH session together at some time next week. I wish to be present when we increase above 20001 Volts for the first few runs... Gun vacuum base is 5.45 E -010 Torr. In normal operations it should not increase above 9.0 E - 010 Torr. See mew during my office hour if you should have any questions. Thank you for your support! James Conway -------------- next part -------------- An HTML attachment was scrubbed... URL: From jhemanth at stanford.edu Wed Jul 19 01:51:16 2006 From: jhemanth at stanford.edu (Hemanth Jagannathan) Date: Wed, 19 Jul 2006 01:51:16 -0700 Subject: Raith free Wednesday 7/19 - 6-9pm Message-ID: <000001c6ab10$7f68aa30$0100000a@jhemanthibm1> Will not be able to use my time. Have removed my reservation from coral -Hemanth -------------- next part -------------- An HTML attachment was scrubbed... URL: From yjhuo at stanford.edu Thu Jul 20 00:58:40 2006 From: yjhuo at stanford.edu (Yijie Huo) Date: Thu, 20 Jul 2006 00:58:40 -0700 Subject: Raith Free Friday 8pm to 1am Message-ID: <004801c6abd2$508f2c80$648a0c80@YijieHuo> Sample is not ready. Best, Yijie -------------- next part -------------- An HTML attachment was scrubbed... URL: From fjaeckel at stanford.edu Thu Jul 20 10:34:19 2006 From: fjaeckel at stanford.edu (Frank Jaeckel) Date: Thu, 20 Jul 2006 10:34:19 -0700 Subject: early finish raith free from 11.45 AM Message-ID: <1153416859.44bfbe9bbdfde@webmail.stanford.edu> started early, will finish early, raith free from 11.45AM - 1PM -- Dr. rer. nat. Frank J?ckel mail address: Department of Chemistry, M/C 5080 333 Campus Drive mailbox #74 Stanford University, Stanford, CA 94305-5080 street address: 369 North-South Mall, Stauffer I, Room 15 Stanford University, Stanford, CA 94305-5080 phone: +1-650-724-4052/51 (office/lab) fax: +1-650-725-0259 (indicate mailbox #74) e-mail: fjaeckel at stanford.edu and frank_jaeckel at web.de From xinranw at stanford.edu Thu Jul 20 11:52:17 2006 From: xinranw at stanford.edu (Xinran Wang) Date: Thu, 20 Jul 2006 11:52:17 -0700 Subject: reservation removed tomorrow from 13:30 Message-ID: <22ffcd060607201152o59cfc347q26a4559a115aa3f5@mail.gmail.com> sample won't be ready. Xinran -------------- next part -------------- An HTML attachment was scrubbed... URL: From jhemanth at stanford.edu Thu Jul 20 18:27:36 2006 From: jhemanth at stanford.edu (Hemanth Jagannathan) Date: Thu, 20 Jul 2006 18:27:36 -0700 Subject: Raith free today from 7pm to 10:30pm Message-ID: <000201c6ac64$d994c2f0$0100000a@jhemanthibm1> Sorry for the late notice. Just found that my samples were not good for the e-beam write Hemanth -------------- next part -------------- An HTML attachment was scrubbed... URL: From jwc at snf.stanford.edu Tue Jul 25 11:56:35 2006 From: jwc at snf.stanford.edu (James Conway) Date: Tue, 25 Jul 2006 11:56:35 -0700 Subject: README: ALL PMMA materials are not to leave the SNF Lab areas. Please be conservative with your usage and help control cost at SNF! Message-ID: <44C66963.8030001@snf.stanford.edu> Greetings Ebeam Lab Users: This is my second message on this thread. [Previous Message in SNF Ebeam archives ] If this situation does not correct itself, I will take action as necessary including locking up all Ebeam Lab polymers and PMMA materials, henceforth forcing ALL Users to check out all volumes of material individually and at greater cost to all Users. ALL Users are instructed to not take from the Cleanroom any Ebeam Polymer or PMMA, nor make bottles of material for their individual usage. If you need a custom viscosities prepared of any type of PMMA -- please see James Conway during his office hour. I have many types already made up and qualified for usage. During July we have gone through 550 ml. of 5% 495K PMMA in Anisole material since July 5, 2006. (Two Bottles of PMMA have been provided to the Clean room on 7-5-06 and 7-13-06) This is an unacceptable amount of usage and is enough material to cover over 300 100 mm wafers with 300 nm of resist thickness on each wafer. Today I stocked a third 250 ml bottle of 5% 495K PMMA in Anisole for use by All Lab Members. I expect all users to utilize these supplies frugally, dispensing all PMMA blends manually through PE/PPVsyringes and 0.45 um Whatman filters only. No direct pouring from the bottles, or into beakers, nor directly onto the SVG wafer spin tracks is allowed whatsoever. The 9% 950K PMMA you may wish to not utilize the filters to reduce your effort at dispense. Please assist us in reducing and holding down the cost of Engineering Materials available for all Lab members working within the SNF Cleanroom! Thank you for your support! James Conway Ebeam Lab -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: manshaking$$$.jpg Type: image/jpeg Size: 26201 bytes Desc: not available URL: From xinranw at stanford.edu Tue Jul 25 21:22:00 2006 From: xinranw at stanford.edu (Xinran Wang) Date: Tue, 25 Jul 2006 21:22:00 -0700 Subject: reservation cancelled friday midnight to 5am Message-ID: <22ffcd060607252122v776e9375hea3119aaa420c109@mail.gmail.com> Was able to finish the writing today, so don't need the time anymore. Xinran -------------- next part -------------- An HTML attachment was scrubbed... URL: From zhangy at stanford.edu Wed Jul 26 22:58:22 2006 From: zhangy at stanford.edu (Yuan Zhang) Date: Wed, 26 Jul 2006 22:58:22 -0700 Subject: Raith free from 6-11pm tomorrow (07/27) Message-ID: <1153979902.44c855fec1bc7@webmail.stanford.edu> Samples not ready. Sorry for the late notice. Yuan Yuan Zhang EE department From jwc at snf.stanford.edu Fri Jul 28 13:19:57 2006 From: jwc at snf.stanford.edu (James Conway) Date: Fri, 28 Jul 2006 13:19:57 -0700 Subject: [Fwd: Comment raith SNF 2006-07-28 11:18:11: 30 kV test completed Users can use any ACCELERATION Voltage 0.3 to 30 KV] Message-ID: <44CA716D.4030506@snf.stanford.edu> Greetings Raith Community: Yesterday we tested for the second time an entire afternoon of write operations at 30 kV with no problems. Gun pressure base level is 5.2 E -010 Torr and system pressure base is 1.9 E - 006 Torr. Very stable gun emission and nominal beam currents levels have been observed. Users can now utilize any ACCELERATION Voltage from 0.3 to 30 KV. ALL SAMPLES GOING INTO THE SYSTEM MUST BE SOFT BAKED JUST BEFORE LOADING INTO RAITH. Thank you for your support, James Conway -------------- next part -------------- An HTML attachment was scrubbed... URL: From ag254 at stanford.edu Sat Jul 29 16:09:52 2006 From: ag254 at stanford.edu (Andrei Gustavo Fidelis Garcia) Date: Sat, 29 Jul 2006 16:09:52 -0700 Subject: RAITH Emergency Error Number 5 Message-ID: <1154214592.44cbeac08101a@webmail.stanford.edu> To whom it may concern: While attempting to load the RAITH I encountered a RAITH error 5. After consulting with a number of RAITH champions, I was unable to resolve the problem. As of right now, the RAITH cannot load or unload any samples. Perhaps there is someone out there who does know how to deal with this error so I have yellow lighted the RAITH on Coral (not red). I apologize to whomever may be affected by this. Andrei Garcia From jhemanth at stanford.edu Sat Jul 29 19:26:31 2006 From: jhemanth at stanford.edu (Hemanth Jagannathan) Date: Sat, 29 Jul 2006 19:26:31 -0700 Subject: RAITH Emergency Error Number 5 In-Reply-To: <1154214592.44cbeac08101a@webmail.stanford.edu> References: <1154214592.44cbeac08101a@webmail.stanford.edu> Message-ID: <1154226391.44cc18d73921b@webmail.stanford.edu> Hi Everyone, I am currently enabled on Raith and am attempting to resolve the problem. I will send out an e-mail to the group once Raith is back up. Best regards, Hemanth Jagannathan Quoting Andrei Gustavo Fidelis Garcia : > To whom it may concern: > > While attempting to load the RAITH I encountered a RAITH error 5. After > consulting with a number of RAITH champions, I was unable to resolve the > problem. As of right now, the RAITH cannot load or unload any samples. > Perhaps there is someone out there who does know how to deal with this > error so I have yellow lighted the RAITH on Coral (not red). I apologize > to > whomever may be affected by this. > > Andrei Garcia > From jhemanth at stanford.edu Sat Jul 29 21:58:35 2006 From: jhemanth at stanford.edu (Hemanth Jagannathan) Date: Sat, 29 Jul 2006 21:58:35 -0700 Subject: RAITH Emergency Error Number 5 In-Reply-To: <1154226391.44cc18d73921b@webmail.stanford.edu> References: <1154214592.44cbeac08101a@webmail.stanford.edu> <1154226391.44cc18d73921b@webmail.stanford.edu> Message-ID: <1154235515.44cc3c7b7edaf@webmail.stanford.edu> Hi Everyone, Arvind and I have cleared the error and have removed the sample from the raith. I did a test write and all seems to be well. The Raith is back on-line. Please ensure that you hold the load-lock door when loading and unloading your samples and if you experience a load lock error please do not do anything innovative unless you know what you are doing. Thanks, Hemanth Quoting Hemanth Jagannathan : > Hi Everyone, > > I am currently enabled on Raith and am attempting to resolve the > problem. I will send out an e-mail to the group once Raith is back up. > > Best regards, > > Hemanth Jagannathan > > Quoting Andrei Gustavo Fidelis Garcia : > > > To whom it may concern: > > > > While attempting to load the RAITH I encountered a RAITH error 5. After > > consulting with a number of RAITH champions, I was unable to resolve > the > > problem. As of right now, the RAITH cannot load or unload any samples. > > Perhaps there is someone out there who does know how to deal with this > > error so I have yellow lighted the RAITH on Coral (not red). I > apologize > > to > > whomever may be affected by this. > > > > Andrei Garcia > > > > > From fjaeckel at stanford.edu Sun Jul 30 15:34:42 2006 From: fjaeckel at stanford.edu (Frank Jaeckel) Date: Sun, 30 Jul 2006 15:34:42 -0700 Subject: early start, free from 1/4 to 5 Message-ID: <1154298882.44cd34023d870@webmail.stanford.edu> -- Dr. rer. nat. Frank J?ckel mail address: Department of Chemistry, M/C 5080 333 Campus Drive mailbox #74 Stanford University, Stanford, CA 94305-5080 street address: 369 North-South Mall, Stauffer I, Room 15 Stanford University, Stanford, CA 94305-5080 phone: +1-650-724-4052/51 (office/lab) fax: +1-650-725-0259 (indicate mailbox #74) e-mail: fjaeckel at stanford.edu and frank_jaeckel at web.de From fjaeckel at stanford.edu Mon Jul 31 12:46:07 2006 From: fjaeckel at stanford.edu (Frank Jaeckel) Date: Mon, 31 Jul 2006 12:46:07 -0700 Subject: raith free from 1PM Message-ID: <1154375167.44ce5dff652d5@webmail.stanford.edu> -- Dr. rer. nat. Frank J?ckel mail address: Department of Chemistry, M/C 5080 333 Campus Drive mailbox #74 Stanford University, Stanford, CA 94305-5080 street address: 369 North-South Mall, Stauffer I, Room 15 Stanford University, Stanford, CA 94305-5080 phone: +1-650-724-4052/51 (office/lab) fax: +1-650-725-0259 (indicate mailbox #74) e-mail: fjaeckel at stanford.edu and frank_jaeckel at web.de From rmstolt at gmail.com Mon Jul 31 12:52:56 2006 From: rmstolt at gmail.com (Randy Stoltenberg) Date: Mon, 31 Jul 2006 12:52:56 -0700 Subject: raith free from 1PM In-Reply-To: <1154375167.44ce5dff652d5@webmail.stanford.edu> References: <1154375167.44ce5dff652d5@webmail.stanford.edu> Message-ID: <8769af0c0607311252g48cca48fn16d89af56384f51d@mail.gmail.com> I'll take it On 7/31/06, Frank Jaeckel wrote: > > > -- > Dr. rer. nat. Frank J?ckel > mail address: Department of Chemistry, M/C 5080 > 333 Campus Drive > mailbox #74 > Stanford University, Stanford, CA 94305-5080 > street address: 369 North-South Mall, Stauffer I, Room 15 > Stanford University, Stanford, CA 94305-5080 > phone: +1-650-724-4052/51 (office/lab) > fax: +1-650-725-0259 (indicate mailbox #74) > e-mail: fjaeckel at stanford.edu and frank_jaeckel at web.de > -- Randy Stoltenberg Ph.D. Student, Department of Chemistry Stanford University Stauffer III, Room 15 381 North South Mall Stanford, CA 94305 randalls at stanford.edu Phone: 650 725 5403 Cell: 650 796 4527 Fax: 650-723-9780