Good Evening,

Users are asked to be vigilant and careful in operation of the RAITH 
system overnight.

Immediately call me if you:

    * Smell hot electrical or burning PCBoards or see smoke.
    * encounter and EHT trip at 10 kV.
    * Encounter other unusual behavior on the system.

Users will observe that the emission current value when measuring Beam 
Currents in the Faraday cups are significantly lower than with the 
former FE-Gun.
Emission should improve as we run the tip in and after we make 
adjustments to the FE-Gun settings with Zeiss Field service.
Emission at 10 kV 30 um aperture is 0.128 nA...

Good Evening,

James Conway

FYI: I have email back up to the jwc at snf.stanford.edu email account and 
all users should send all email to that default address.  Thanks. JWC

-------- Original Message --------
Subject: 	Comment raith SNF 2006-10-19 18:25:29: system is up and 
functional NOT TO EXCEED 10 KV
Date: 	Thu, 19 Oct 2006 18:25:30 -0700
From: 	jwc at snf.stanford.edu
To: 	raith-pcs at snf.stanford.edu



Greetings:
All systems and modules are functional.
Users are not to go above 10 keV acceleration voltage as we had an EHT TRIP during the first test at 15 kV.
Vacuum remains good at 5.8E-010T for gun and 1.6E-006 for chamber.
Good Night,
JWC