Problem raith SNF 2007-08-25 16:48:05: No TV / Inlens image and how to recover...
James Conway
jwc at snf.stanford.edu
Tue Aug 28 10:43:24 PDT 2007
Greetings Raith Users:
We have seen this issue surface several times through the course of the
last week and rarely in the past.
This is normally a result of the L-REM SEM control module on the SEM
computer not registering with the LEO server upon user login and
application launch.
To recover from this issue you should gracefully exit the application
and shut down the Windows 2000 OS, allow ten seconds after the SEM
computer screens turns black, and then press the front panel power
button to power up the SEM PC. The RAITH PC and the RAITH150.exe
application DO NOT have to be shut down or exited. But be sure that REM
CON on the SEM computer has its communication channel open to allow
communications between the computers.
If this does not repair the problems please contact James Conway. You
can also refer to FAQ No. 3: How to Reinstall Detectors to the LEO
Configuration File" listed in the beginning of the RAITH 150 Notebook
No. Three. You will need to obtain a password to access this
configuration page.
I am continuing to evaluate this problem closely to determine the root
cause of this intermittent problem. Users are requested to call me
immediately if this occurs during your day time RAITH session.
Thank you,
James Conway
koseki at snf.stanford.edu wrote:
> No TV / Inlens image are obtained. The starting up log of the LEO-SRV says "L-REM is not responding".
>
>
-------------- next part --------------