Problem RAITH SNF 2007-02-14 16:21:21: Standard sample holder not landing onto piezo post properly
James Conway
jwc at snf.stanford.edu
Wed Feb 14 19:18:34 PST 2007
kevin or cole,
Hope to speak with you in the morning tomorrow.
I mapped both wafer height on the Std. Sample holder and the surface of
the sample holder itself after driving the center of the wafer (0,0 UV)
to 5 mm and see significant offsets from level despite using a
calibrated center positions of the piezo post. Up to 0.5 mm across the
pallet. I made some maps of heights from two different loads on the
Std. Sample Holder.
Following up with phone with Kevin last Friday:
Yes I think we have a problem with P1 the fixed piezo post.
I have been troubleshooting some strange reports from users whom
indicated that initially the Z drive was loosing its calibration; when
in fact the sample holder may not be landing correctly onto the P mounts
or now more likely the jewel tip on P1 may have become damaged or
simply worn itself out.
Also have been troubleshooting several reports of strange EBL results
such as triangles and 100 nm squares being circular and not well defined
. I suspected out of focus by users now I suspect the P1 Post.
Flatness of the Std. sample holder appears fine when placed on machine
plate gage. i.e., it doesn't rock or appear bent. But I don't have any
dimensions or specs for this assembly to check to.
I have some SEM's of each post... P1 looks really different that the
others. cracked or worn on the periphery.
Problem RAITH SNF 2007-02-14 16:21:21: SUBJECT: Standard sample holder
not landing onto piezo post properly.
Attention All RAITH Users:
PROBLEM ON RAITH: STANDARD SAMPLE HOLDER NOT LANDING ON KINETIC MOUNTS
-- DAMAGED PIEZO POST...
I have been troubleshooting several user reports of irregularly shaped
or poorly defined written structures on the exposures made since January
11, 2007. This issue has become more apparent through the month, and as
we perform hi resolution measurements of our structures in SEM review
written in early January.
I have observed several incidents of out of focus or poorly focused
structures written at the extreme edges of wafers ( E.g., targets and
SPLines @ 0, -45 UV :: 0 +45 UV) Within the normal +- 10 mm range I
often write within, I also observed significant changes in working
distance as I moved through this small range of positions.
Monday and today we made measurements of the surface of both a 100 mm
wafer clipped to the standard sample holder and upon the Std. sample
holder itself and have collected data that the holder is not landing
precisely flat into the kinetic mounts. The range of non-flatness is ~
0.500 mm!
Visual and SEM inspection of the ceramic balls and points or 'jewel'
tips in the Piezo post yield information that indicated that Piezo one
post ceramic ball may have become damaged and/or the point may be
missing. All piezo post exhibit wear on their surfaces.
Users are requested to be vigilant of their actual working distances,
and to report their WD data for specific positions across their samples
and promptly report to this list if you see significant changes in WD as
you move across your wafers greater than 250 um.
RAITH has been contacted last Friday and we will strive to correct this
issue ASAP.
System is now in YELLOW Light Mode.
Thank you for your support!
James Conway
More information about the raith