James Conway jwc at snf.stanford.edu
Fri Oct 19 17:36:36 PDT 2007


Greetings Raith Community,

I have made a best effort to resolve the cooling pump issues primarily 
related to the cooling jacket in the Turbo pump becoming clogged with 
scale and particulate debris.
I purged the system of bubbles and checked for leaks and all looks 
good.  While flow initially was pegged at over 2.8 liters per minute it 
has fallen to the original value of 1.6 lpm we obtained shortly after 
the last system flush several months ago.  It is my expectation that 
this will get us through some period of time; but likely did not address 
all issues related to the RAITH system cooling loop.
I am going to schedule downtime for maintenance with RAITH field service 
at the earliest opportunity.

Users will note that emission levels are slightly below where they were 
two days ago -- but this will steadily increase as Gun section vacuum 
decreases and emission levels stabilize to optimal values.
I/O and E/O functions have been tested and Users will not have any 
problems they encountered Tuesday evening if they carefully follow the 
Load Lock procedures. 
BE SURE TO HOLD THE Load Lock DOOR DOWN!

I re-homed all positions for the LIStage and updated the recorded 
positions of the Faraday cups back to the Std. Sample Holder defaults.
I tested the SEM functions and have performed an EBL write that is 
exposing as I write this email.

Users should attempt to complete as much EBL work as possible before we 
take the system down for more repairs and maintenance at some point next 
week.
RAITH Fields Service has been requested to come on site as soon as possible.

Special Thanks to Dmitry Kozak and the other users whom assisted me 
during this difficult period.

Thank you for your support!

James Conway




-------- Original Message --------
Subject: 	Re: Problem raith SNF 2007-10-19 16:04:52: FE-GUN run up in 
progress.
Date: 	Fri, 19 Oct 2007 17:05:48 -0700
From: 	jwc at snf.stanford.edu
To: 	raith-pcs at snf.stanford.edu



full system check out completed.
preliminary qualification write in progress.
FE-Gun will take a day or so to reach deep vacuum and optimal emission levels. 
Ready for next user after 18:00 hours
turbo pump and cooling loop will need flushing and exchange next week. Raith field service requested... 
JWC