Problem with PMMA exposure/lift off
James Conway
jwc at snf.stanford.edu
Thu Apr 10 11:01:16 PDT 2008
Hello Arash Hazeghi and Others:
Thanks for the images.
In the first two images you are below the dose to clear for the finer
lines. What is evident is you can see the 1 um sub-sub-field butting as
these field overlap slightly, and at the 40 um sub-field you can see an
apparent stitch break between the write fields. It is normal practice
to increase dose slightly to eliminate these effects and both of those
exposure artifacts will develop out. If your resist stack height is
above 400 nm you may wish to increase development time and agitation as
well. On the 300 nm 5% 495K PMMA I routinely develop fro 30 seconds,
perform a quick high Magnification Microscope inspection, and then if I
see these artifacts perform additional development in ten second
intervals until the features are cleared.
As the EBEAM tool slowly accumulates hydrocarbons in the lower sections
of the column, specifically the final aperture in the objective lens and
in the area around the blanking aperture, the astigmatism in the lens
increases as does some charge up in the apertures when blanked;
resulting in shifts of the ebeam spots placement as it writes these
sub-fields. Within a very small range these fields can be adjusted --
but at some point we may need to perform a PM on the column to return
the system to a normal state. I am currently evaluating these effects
and have just two evenings ago refocused and changes stigmation slightly
to the column settings. In my exposure test I reviewed last evening I
reported a reduction in the magnitude of the stitching issues
particularly for the 40 um sub-fields.
IN the last two images it is plainly evident that there is an adhesion
issue in your metal lift off films. You may wish to reconsider your
initial substrate clean up method to ensure you are starting out clean
and without any small source residues on your wafers surface. A simple
TAPE test will assist you in evaluating thin film adhesion.
All Users are advised to be sure to run the $DEF CF CL write field
calibration routine for all of their exposures. Users working on any
charging substrates such as SOI, or substrates with oxides or nitrides
are recommended to coat their samples with a charge compensation coating
such as ESPACER or a thin metal film to reduce charging effects on your
EBL writes.
Thank you for your support!
James Conway
Arash Hazeghi wrote:
> Hi all,
> I have been trying liftoff on my samples recently but it looks like
> either I am doing something wrong or there is an exposure problem with
> eBeam. I am trying to resolve 150nm lines on 300nm 5% 950-K PMMA
> (standard coat procedure). I have tried area exposure dose from
> 350uC/cm2 to 430uC/cm2. metal is 80A Pd+200A Pt. lift off was in
> acetone with syringing. looking under the optical microscope some
> higher dose exposures seem to have come out except for there are cuts
> in the thin lines. In other samples looks like the exposed areas were
> also lifted off. I have attached several pictures, exposure does is
> indicated in each filename. Some users have already suggested using
> higher doses which I am trying now, but I'd be happy to hear if there
> are any other comments.
>
> Thank you,
> Arash
>
>
>
>
>
> --------------------------------------------------------------------------
> Arash Hazeghi
>
> PhD Candidate
> Stanford Center for Integrated Systems,
> 420 Via Palou Mall, CIS-X 300
> tel: 650-725-0418
> mobile: 650-353-1866
> http://www.stanford.edu/~ahazeghi <http://www.stanford.edu/%7Eahazeghi>
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20080410/0bc3ab32/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: not available
Type: image/jpeg
Size: 154072 bytes
Desc: not available
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20080410/0bc3ab32/attachment.jpe>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: not available
Type: image/jpeg
Size: 154752 bytes
Desc: not available
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20080410/0bc3ab32/attachment-0001.jpe>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: not available
Type: image/jpeg
Size: 156748 bytes
Desc: not available
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20080410/0bc3ab32/attachment-0002.jpe>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: not available
Type: image/jpeg
Size: 151241 bytes
Desc: not available
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20080410/0bc3ab32/attachment-0003.jpe>
More information about the raith
mailing list