This afternoon Lev Markov from Raith USA came in for some additional
work on the problematic turbo cooling situation. In the process of
investigating and localizing flow limitations in the system, a kink in
one of the water lines was found. Once the kink in the line was
straightened, a healthy flow of about 2.3 liters per minute at a chiller
pressure of only about 40 PSI was observed. Yesterday at this time, it
took 80 PSI to force only 1.2 liters per minute through the system.
While we will continue to monitor the flows, we are optimistic that this
may resolve this problematic issue.