Good Afternoon Byoungil Lee
Great Question -- soon to be a new FAQ!
Smaller step sizes will give you an improvement in Line Edge roughness
for high contrast resist but should not necessarily affect ultimate
writing resolution.
Yes it will increase the beam speed if the dwell time is the same....
If you need a smaller shot density consider using a smaller write field
size so you can place more shots within the same area dimension. Shot
grid, not to be confused with the 1 nm design grid, is the write field
dimension divided by 2n**16 (Two n to the 16th)
With either higher beam speeds or lower settling time you will start to
see dynamic effects in your EBL writes. Particularly evident in Y
deflections EBL writes may appear chirped or distorted compared to the
pattern design. For Example, a cross bar target might have little tails
on the beginning or end of the X or Y cross bar features. There may
also be orthogonality distortions evident.
Best beam speed is between 4 and 7 mm per second and you will never see
any dynamic effects in normal exposures with proper WF alignment.
10 mm/sec is the highest recommended beam speed for quality writes.
10 to 15 mm/sec will likely not show defects in normal writes but more
difficult patterns such as Single Pixel lines may have some odd
beginning and end features or breaks.
Once you exceed 25 mm / second it all falls apart and a host of defect
deflection and exposure artifacts may present themselves.
To decrease your beam speed for the same dose to clear you can increase
your area step, or reduce your total Beam Current (I sub beam) by
selecting the next smaller aperture size.
here is one example:
Inner part of a Global Mark in the demo.csf pattern written in 100 nm
PMMA with a 0.00156 area step versus a 0.016 um area step. The beam
speed was ~ 25 mm/sec. Intra pattern break evident as well as blocks of
the pattern data wrote out of normal alignment. There are NO write
field breaks in this pattern....
Hope this answers your questions well.
James
Byoungil Lee wrote:
> Hi James,
>
> I have a question regarding the step size and beam speed for the
> exposure parameters in Raith.
> I know that smaller step size is better for smaller patterns(better
> resolution from smaller step size). But, smaller step size increases
> the beam speed.
> If the beam speed is too high (higher than 10mm/s), would that cause a
> problem? If so, what kind of problem will be caused by high beam speed?
> Also, what would be the optimal step size and beam speed for 40-100nm
> width lines?
>
> Thank you!!
>
> -Byoungil
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20090501/8c32d22d/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: MARK1.BMP
Type: image/bmp
Size: 91078 bytes
Desc: not available
URL: <http://snf.stanford.edu/pipermail/raith/attachments/20090501/8c32d22d/attachment.bmp>
More information about the raith
mailing list