Comment raith SNF 2011-12-03 16:40:20: Update on External Scan Control
James W. Conway
jwc at snf.stanford.edu
Wed Dec 7 18:19:40 PST 2011
OK Thanks everyone for your reports!
I spoke with Kevin Burcham yesterday during my conference call and he
indicated that it should not matter either way -- but that most systems
have reverted to launching the SMARTSEM first before the RAITH
application launch.
EVERYONE USING RAITH 150:
Let change the start-up procedure to start with the launch of the
SMARTSEM package, log in on the SMARTSEM client, and then launch the
RAITH 150 application. Same as we used to do.
I will be testing this when I am next on the tool and with user sessions
tomorrow before shutdown.
Thank you
James
On 12/5/2011 5:50 PM, Matthias Baenninger wrote:
> Hi James,
>
> I had that problem with the SEM not switching to external control again
> during my write on Friday. I used the same order of starting up as Kevin
> describes, i.e. I started Raith150 and the SmartSEM. When switching to
> external control manually, everything worked fine.
>
> Matthias
>
>
> -----Original Message-----
> From: James W. Conway [mailto:jwc at snf.stanford.edu]
> Sent: Monday, December 05, 2011 12:17 PM
> To: khu834 at snf.stanford.edu; Ebeam Lab
> Subject: Re: Comment raith SNF 2011-12-03 16:40:20: Update on External Scan
> Control
>
> Kevin,
>
> Thanks for this observation. I will continue to test to see what trips
> this.
>
> It shouldn't matter either way as multiple hand shakes in communications
> occur before I get to the writing part of the sequence.
>
> If other Users are seeing the horizontal streaks in their patterning,
> missing patterns or strange exposure artifacts resulting from external scan
> control not switching during your patterning -- please report back to this
> list. I am tracking all exposure errors on the system. I think it is only
> a rare intermittent problem at this point....
>
> James
>
> On 12/3/2011 4:40 PM, khu834 at snf.stanford.edu wrote:
>> It seems that this problem only occurs when lauching Raith first, followed
> by the SmartSEM.
>> If you open SmartSEM, then Raith, everything functions as normal
>>
More information about the raith