Problem sem4160 2003-07-28 18:38:39: Report Clear for sem4160
charley at snf.stanford.edu
charley at snf.stanford.edu
Tue Jul 29 02:23:36 PDT 2003
Found emission set to 5uA: wrong! Found condenser
set to 10, also wrong. If you lower the emission, what's the
point of opening the condenser. DO NOT change emission.
This is 10uA and when set accordingly, I had no trouble
running for an hour. No problems. Up for use. Use about 12
for condenser. See me if you have trouble, before changing
the settings, please!