Problem sem4160 SNF 2007-08-08 00:45:44: Sample Holder

James Conway jwc at
Wed Aug 8 09:55:00 PDT 2007

Be sure the loading rod is pulled out fully out to its limits in the 
load lock chamber. There is a touchy micro-switch that needs to make 
before the LOAD Lock exchange door valve will open on switch actuation.

See me if you have more problems


nppatil at wrote:
> Cannot unload sample holder from Specimen Chamber to Specimen Exchange Chamber.