Greetings /semhitachi /and /sem4160 /Users:
*Users are requested to not translate the upper movable aperture on
either of the SEM columns.*
This is called the Beam Monitor Control aperture and its function is to
monitor the beam current at that aperture during SEM operations on the
tools. Often Users will see the BMC button flashing as emission of the
FE-Tip falls below a specific set point (~ -20% delta Emission
Current). You can bring the emission back up to the 10 µA set point
simply by pressing the 'HV ON' button again without shutting the beam down.
The lower movable aperture strip called the Beam Alignment Aperture, it
can be translated by any User to align the Beam to the center of the
aperture resulting in no image shift as you change working distance (focus).
If you are unsure, confused, or just need help, please see James Conway
during any training session on these tools, or visit me during my office
hour. Tuesday - Friday 8:30 - 9:30 AM daily.