Hello Matt and /semhitachi /and/ sem4160/ Users:

Both of our SEMs are calibrated to a NIST traceable line width standard 
using  Geller MRS3 ebeam written calibration chip.  The value you 
measure on either SEM is within 0.5 % in X plane and within +- 1.0 or 
less in Y plane.  This calibration is checked every other month at 100, 
1000, and 100,000X magnification.

Please see me if yo have questions or would like to join me for a 
calibration and metrology session with me.

Thank you,


Matthew W. Messana wrote:
> Has anyone compared the scale bar on semhitachi images to a known sample dimension? I'm looking to measure a 5um film and am curious whether or not I can trust the scale bar.
> Thanks,
> Matt