There are two users who are anxious for a SpecMat decision. They are
fairly simple questions, which I was hoping we could resolved through email
before next week's meeting. Please comment on both requests.
1) Rhett Brewer:
A) Can I use rtaag to do a ~250C O2 anneal of my Pd, Si, O film with ~3-4
monlayers worth of carbon.
B) Want to verify that my film stack: si(100)/Si,Pd,O/200A Al2O3/5000A
P-doped poly Si is allowed in the RTA 4108. Want to do a 2 min., 650C N2
or O2 anneal.
Ed's Comments:
A) Should be fine in RTAag since it runs under thermal couple control.
B) The Si-Pd-O film will be buried under our polysilicon film. This
should be fine as long as he undergoes the standard pre-RTA clean and use
the silicide tray.
2) Carl Faulkner
We would like to use Tylan4 - not for forming gas, but for 800C - 1000C
anneals in nitrogen. The samples would be Ti- or W-capped LaB6 on
otherwise standard Si/oxide/nitride substrates. LaB6 has a melting point >
2000C.
Ed's Comments:
I don't see a problem with this request as long as there is an appropriate
clean. Tylan4 is no longer considered a clean furnace.
Regards,
Ed